Dear Holger/Sabine/ all.
We looked at Anton-Paar's rolling ball viscometer about 7 years ago and found that silanization of the capillary led to considerable improvement: Anton-Paar - who were really helpful - did this for us on request and I think they still do this, although there is a charge.  [nb. We found standard Ostwald viscometers are still considerably more accurate than the rolling ball for dilute solution intrinsic viscosity work].
My student Yanling Lu wrote up a section on this in her PhD thesis, I will see if I can find it and send it on,
All best
Steve

**********************************************************
Stephen E. Harding, http://www.nottingham.ac.uk/Biosciences/People/steve.harding
Professor of Applied Biochemistry, University of Nottingham
E-post: steve.harding at nottingham.ac.uk<mailto:steve.harding at nottingham.ac.uk>
Tel: +44(0) 78110 90635

From: rasmb-bounces at list.rasmb.org [mailto:rasmb-bounces at list.rasmb.org] On Behalf Of Sabine Kaltofen
Sent: 19 August 2013 13:02
To: HGSR (Holger Martin Strauss); rasmb at list.rasmb.org
Subject: Re: [RASMB] capillary viscosimeter

Hi Holger and all,

have you thought of coating the capillary with a silanizing agant such as SigmaCote that is used in crystallography? It works fine for quartz cuvettes and forms a fairly stable film (you have to boil it in detergent to remove it). I guess you have to recalibrate your capillary, though.

Best
Sabine

Mind your carbon footprint: please don't print this message.


Sabine Kaltofen

Biochemistry & Structural Biology
Lund University
+46 462220851
sabine.kaltofen at biochemistry.lu.se<mailto:sabine.kaltofen at biochemistry.lu.se>
________________________________
From: rasmb-bounces at list.rasmb.org [rasmb-bounces at list.rasmb.org] on behalf of HGSR (Holger Martin Strauss) [hgsr at novonordisk.com]
Sent: 19 August 2013 10:46
To: rasmb at list.rasmb.org
Subject: [RASMB] capillary viscosimeter
Hello everyone,

We've recently ventured into measuring viscosities. We use a capillary viscosimeter Lovis 2000 ME from Anton Paar and have observed several times that high concentration protein solutions stick to the (glass) capillary, after rinsing and scrubbing. We then treat the capillary with piranha solution. This is annoying and cumbersome and we haven't yet found a way to avoid/reduce of protein sticking to the surface. But maybe someone of you has?

Best wishes, Holger

_______________________________

Holger Martin Strauss, PhD.
Senior Scientist
Biophysics

Novo Nordisk A/S
Novo Nordisk Park
DK-2760 Måløv
Denmark
+45 4444 8888 (phone)
­+45 30 75 66 94 (mobile)
hgsr at novonordisk.com<mailto:hgsr at novonordisk.com>

Changing diabetes since 1923
Did you know that Novo Nordisk was the first company to introduce insulin to Scandinavia? Find out more here<http://www.novonordisk.com/about_us/changing-diabetes-activities/NN-history.asp>

This e-mail (including any attachments) is intended for the addressee(s) stated above only and may contain confidential information protected by law. You are hereby notified that any unauthorized reading, disclosure, copying or distribution of this e-mail or use of information contained herein is strictly prohibited and may violate rights to proprietary information. If you are not an intended recipient, please return this e-mail to the sender and delete it immediately hereafter. Thank you.


This message and any attachment are intended solely for the addressee and may contain confidential information. If you have received this message in error, please send it back to me, and immediately delete it.   Please do not use, copy or disclose the information contained in this message or in any attachment.  Any views or opinions expressed by the author of this email do not necessarily reflect the views of the University of Nottingham.



This message has been checked for viruses but the contents of an attachment

may still contain software viruses which could damage your computer system, you are advised to perform your own checks. Email communications with the University of Nottingham may be monitored as permitted by UK legislation.









-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20130819/146a2aa2/attachment.htm>


More information about the RASMB mailing list