Dear RASMB
Wondering about the s020w calculation in SEDNTERP
For one of our polymer micelles with which we've been working, a SEDANAL
analysis returns an s20w value (not corrected for concentration) of 17.3.
If I use equation 19 in cited in the SEDNTERP documentation, from the s20w
value=17.3, I calculate s* = 16.4s.
Phsyical parameters used:
1. experiment temp=20oC,
2. vbar=0.8312 (measured by densitometry at 20oC)
3. density of solution 1.00506 gms/cm3 (measured)
4. viscosity 0.010281 (measured)
5. Assuming viscosity of water to be 0.01002, and the density to be
0.998234
However if I use the Experiment page in the SEDNTERP and enter value of s*
calculated above, then s20w returned is 27.8 which is much higher than the
value returned from SEDANAL. In fact in order to get SEDNTERP to return a
value of 17.3s, I have to decrease S* to be 2.91s.
In SEDFIT, if I input 16.4s, calculated from Equation 19 in the SEDNTERP
documentation, and input the same physical parameter listed above, I get
17.4s in agreement with the SEDANAL result.
I think in both SEDNTERP and SEDFIT, the same equation is used to correct s*
to s20w, so I'm not clear what I'm missing.
Any suggestions and comments are welcome and thank you, in advance, for
taking the time to consider my question.
Best regards,
John Sumida
<http://depts.washington.edu/cidb4bio/index.shtml> Analytical Biopharmacy
Core Facility
University of Washington
<http://www.moles.washington.edu/> Molecular Engineering & Sciences
Institute, G22
3946 West Stevens Way NE
Seattle WA 98195-1653
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20130808/9635d56c/attachment.htm>
More information about the RASMB
mailing list