[RASMB] Correspondence concentration cesium chloride vs fringes
Antje Voelkel
antje.voelkel at mpikg.mpg.de
Tue Nov 29 06:02:11 PST 2011
Dear Eric,
I think the dn/dc of the CsCl varies from package to package. We ones
measured a value of 0.0837 ml/g.
With the equation:
*delta c pro fringe shift J*
delta c [g/ml] = J * lambda*10-7 / dn/dc * a
dn/dc [ml/g]
a [cm - 1.2 for standard AUC cells]
lambda = 675 nm (laser)
you are able to calculate every concentration you want to.
(Investigation of quantum size colloids using the XL-I
ultracentrifuge;/H. Cölfen
<http://www.springerlink.com/content/?Author=H.+C%c3%b6lfen>//, //T.
Pauck <http://www.springerlink.com/content/?Author=T.+Pauck>//und //M.
Antonietti <http://www.springerlink.com/content/?Author=M.+Antonietti>/)
For the simulation of a gradient, you need activity coefficients. I can
send you an excel sheet for the calculation of the concentration gradient.
The determination of density distributions and density gradients in
binary solutions at equilibrium in the ultracentrifuge/Ifft J.B., Voet
D.H. Vinograd J. J.Phys.Chem 65/1961 1138-1145/
//
Best regards
Antje
/
/
Am 29/11/2011 07:33, schrieb Kusznir, Eric Andre:
>
> Hello RASMB Community
>
> I will be grateful if any of you could tell me the correspondence of
> the gradient concentration of cesium chloride based on the number of
> interference fringes.
>
> Thanks in advance.
>
> Best regards
>
> Eric
>
>
>
> _______________________________________________
> RASMB mailing list
> RASMB at rasmb.bbri.org
> http://rasmb.bbri.org/cgi-bin/mailman/listinfo/rasmb
--
Antje Voelkel
Max Planck Institute of Colloids and Interfaces
Am Muehlenberg 1
D-14476 Potsdam
+49(0)331 567 9528/9531
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20111129/58ac8367/attachment-0001.html>
More information about the RASMB
mailing list