Kristian,
It looks like the radial calibration has been corrupted somehow. You're
getting an infinite value (+INF) for the cm/pixel and you're getting
"not-a-number" (-NAN) for the first radius, and then infinite (+INF) for the
second radial point. I don't remember whether the pixel column values that
are used to calculate the magnification are stored in the cal_parm.cal file
or somewhere else.
Sorry that I can't be more helpful,
Jeff Lary
Biotechnology Center Unit-3149
ph: 860-486-5036
fax: 860-486-5005
email: jeffrey.lary at uconn.edu
===========================
-----Original Message-----
From: rasmb-bounces at server1.bbri.org [mailto:rasmb-bounces at server1.bbri.org]
On Behalf Of Kristian Schilling
Sent: Monday, May 21, 2007 5:43 AM
To: rasmb at server1.bbri.org
Subject: [RASMB] Error in interference scan acquisition
Dear all,
we have encountered a strange error with our interference optics. During our
last experiment, we obtained scans that look as the following:
======== begin of scan file ===========
cm/pixel: +INF, xxx sample name xxx
P 1 20.3 1100 0000005 4.1546E05 675 1
-NAN 0.00000E+00
+INF 5.93972E-03
======== end of scan file =============
As it appears, the data (but for cm/pixel) in the scan header is correct, so
the machine does communicate with the software, and experimental parameters
are transferred correctly. But instead of the actual scan data, two lines
are written that may present some diagnostic information. Has anybody seen
this before and has an idea what these lines tell us?
The situation has somewhat escalated since then as the software cannot be
launched anymore and there seems to be a serious hardware error that is not
due to any of the boards.
The scan above might give us an idea as to the failure preceding the
failure.
Thanks for any input,
Kristian
=================================
Nanolytics
Gesellschaft fuer Kolloidanalytik mbH
Dr. Kristian Schilling
Am Muehlenberg 11
14476 Potsdam
Tel.: +49 331 5818360
Fax: +49 331 5818361
e-mail: schilling at nanolytics.de
Internet: www.nanolytics.de
_______________________________________________
RASMB mailing list
RASMB at rasmb.bbri.org
http://rasmb.bbri.org/mailman/listinfo/rasmb
More information about the RASMB
mailing list