Dear Jeff,
thanks for your input and for taking the trouble to search for the exact
location. I find this explanation highly probable, and it could also
explain why the bc<some number I forgot>rtl.dll crashed when we later tried
to launch the xli software. As a box came up that explicitely stated a
loating point divide by zero. Apparently the dll was attempting some
calculation with corrupted initialization values and was not defaulted to
some reasonable action for the case of a zero in the wrong place...
However, as the service technicians (in the meantime, their number has
increased to two) have bulldozed the windows installation and are
reinstalling the complete system (right now they are caught with driver
problems that may occur when we try to install an eight year old operating
system on an eight year old PC...), we can't check the xla.ini file any
more. Now this would have been really interesting...
Thanks again,
Kristian
At 10:35 21.05.2007 -0400, you wrote:
>Kristian,
>After looking further I've found that the interference calibration numbers
>are stored in the file XLA.INI, located in your \Windows directory. The
>parameters are in a section that looks like the following,
>==============================
>[RadCal]
>InRad=5.850
>OutRad=7.15
>InPixel=141
>OutPixel=1872
>==============================
>
>If your XLA.INI file is corrupt or if the numbers don't contain realistic
>values then that may be causing the problem. If the numbers are simply
>missing then the following default values will be used
>
>InRad=5.850
>OutRad=7.15
>InPixel=0
>OutPixel=2048
>
>Jeff Lary
>Biotechnology Center Unit-3149
>ph: 860-486-5036
>fax: 860-486-5005
>email: jeffrey.lary at uconn.edu
>===========================
>
>
>-----Original Message-----
>From: rasmb-bounces at server1.bbri.org
>[mailto:rasmb-bounces at server1.bbri.org] On Behalf Of Kristian Schilling
>Sent: Monday, May 21, 2007 5:43 AM
>To: rasmb at server1.bbri.org
>Subject: [RASMB] Error in interference scan acquisition
>
>Dear all,
>
>we have encountered a strange error with our interference optics. During
>our last experiment, we obtained scans that look as the following:
>
>======== begin of scan file ===========
>cm/pixel: +INF, xxx sample name xxx
>P 1 20.3 1100 0000005 4.1546E05 675 1
> -NAN 0.00000E+00
> +INF 5.93972E-03
>======== end of scan file =============
>As it appears, the data (but for cm/pixel) in the scan header is correct,
>so the machine does communicate with the software, and experimental
>parameters are transferred correctly. But instead of the actual scan data,
>two lines are written that may present some diagnostic information. Has
>anybody seen this before and has an idea what these lines tell us?
>
>The situation has somewhat escalated since then as the software cannot be
>launched anymore and there seems to be a serious hardware error that is
>not due to any of the boards.
>
>The scan above might give us an idea as to the failure preceding the failure.
>
>Thanks for any input,
>Kristian
>
>
>=================================
>Nanolytics
>Gesellschaft fuer Kolloidanalytik mbH
>Dr. Kristian Schilling
>
>Am Muehlenberg 11
>14476 Potsdam
>Tel.: +49 331 5818360
>Fax: +49 331 5818361
>
>e-mail: schilling at nanolytics.de
>Internet: www.nanolytics.de
>
>
>_______________________________________________
>RASMB mailing list
>RASMB at rasmb.bbri.org
>http://rasmb.bbri.org/mailman/listinfo/rasmb
>
>
>
>__________ NOD32 2281 (20070521) Information __________
>
>Diese E-Mail wurde vom NOD32 antivirus system geprüft
>http://www.nod32.com
More information about the RASMB
mailing list