Mark, I did experiments at low temperature before.  I'm not sure about your
error, but the thing you have to be careful with is that when you go below
100 micron, the temperature that you measure is the one of the rotor and not
the one in the chamber, so it is possible that in order to keep the rotor at
10 C or even to cool it down that low, the temperature in the actual chamber
goes down a lot more than 10C, almost to freezing levels. Similarly, when
you try to heat the rotor to 40C the temperature in the chamber can go up to
60 or 70C. To get around this problem I suggest you cool down your rotor in
a fridge (even at 4C) before you load it into the XLI, that way you'll
stress the system a lot less.

Good luck
Barbara

-----Original Message-----
From: rasmb-bounces at rasmb.bbri.org [mailto:rasmb-bounces at rasmb.bbri.org] On
Behalf Of Mark Agacan
Sent: Thursday, June 14, 2007 8:49 AM
To: rasmb at server1.bbri.org
Subject: [RASMB] Diagnose the error message

Hello,

My XL-I was running overnight at 10 degrees C, at some point in the
experiment an error was received:

DIAGNOSTIC  12948684800 RECEIVED

The entire roll of print-out paper in the XL-I spewed out with the above
error message to form a pile roughly the size of the XL-I itself.  I
checked the troubleshooting section in the manual, an error message
starting with a 1 implies a 'CPU-Microprocessor malfunction or loss of
program memory'-type error.

The XL-I worked fine the next day but I was working at 16 degrees C. 
Could temperature be the cause of the above error message.  We have had
problems with big temperature fluctuations recently, the XL-I motor
seems to generate a lot of heat, with the result that i have to keep the
lab door open to keep the experiment at a stable temperature.  For
example, it took around twice as long (> 5 hours) for the XL-I to fully
thermally equilibrate at 10 C as it did for 16 C.

Does anyone have any knowledge of the above error, or has anyone had
serious issues with the temperature stability of an XL-I?

Cheers,

Mark


_____________________________________
Dr Mark Agacan
Scientific Officer,
Division of Biological Chemistry 
and Drug Discovery,
Wellcome Trust Biocentre,
College of Life Sciences,
Dow St., 
University of Dundee,
Dundee, DD1 5EH
Tel: +44 1382 388751
Fax: +44 1382 345764
_____________________________________
_______________________________________________
RASMB mailing list
RASMB at rasmb.bbri.org
http://rasmb.bbri.org/mailman/listinfo/rasmb




More information about the RASMB mailing list