Christine,
I am wondering if your real problem is a cooling problem rather than a pump
problem. Have your service engineer check that all the cooling fans are
running properly, and that the air outlets are not clogged with dust.
John
-----Original Message-----
From: rasmb-bounces at rasmb.bbri.org [mailto:rasmb-bounces at rasmb.bbri.org] On
Behalf Of Christine.Wandrey at epfl.ch
Sent: Sunday, June 25, 2006 2:34 AM
To: rasmb at rasmb.bbri.org
Subject: [RASMB] Vacuum pump
Dear all,
two weeks ago, after more than 8 years, the
vacuum pump in our XL-I died (the pump became
very noisy).
The Beckman service first changed the motor.
Though the first run after this reparation
stopped after about 4 hours with the message that
there was no vacuum after running about 4 h at 40
000 rpm.
Then the service recommended to replace the old
pump by a new one. We agreed and the service
arrived with a new one. This one became very hot
already during the test run. The service had to
come again with a another new pump. With the new
pump the 0 microns were reached after 20 nim.
However, after the departure of the service, the
subsequent run stopped after about 2 hours with
the same message: bad vacuum.
The service came again now running the XL-I and
measuring the temperature in the pump. After
about 4 hours the pump/oil temperature approached
75°C and stopped, with the consequence that the
AUC stopped running too.
Having identified the problem there is my
questing: Is there a special pump type that I
should request from the service in order to avoid
this problem? Has anybody similar experience?
Thank you in advance for your recommendations.
With kind regards,
Christine
--
PD Dr.habil. Christine Wandrey, MER
Ecole Polytechnique Fédérale de Lausanne (EPFL) EPFL-SB-ISIC-LMRP2, station
15 CH-1015 Lausanne Suisse
tél: (+41) (21) 693 36 72 in CH J2 494, Station 6
or (+41) (21) 693 96 61 in AAB 042, Station 15
fax: (+41) (21) 693 60 30