With regard to Chad's question about the accuracy of the calculated vbar, I
have seen and published one case where the calculated vbar was clearly way
off, giving an apparent MW 9% lower than the sequence value. The
Edelstein-Schachman D20/H20 method proved the calculated vbar was wrong. See
Philo, J. S., Yang, T. H., and LaBarre, M. (2004). Re-examining the
oligomerization state of macrophage migration inhibitory factor (MIF) in
solution. Biophys. Chem. 108, 77-87.
John
-----Original Message-----
From: rasmb-bounces at rasmb.bbri.org [mailto:rasmb-bounces at rasmb.bbri.org] On
Behalf Of Chad Brautigam
Sent: Tuesday, June 06, 2006 11:46 AM
To: rasmb at rasmb.bbri.org
Subject: [RASMB] vbar
Hello, All,
Sorry if this is a rudimentary question. I have been using velocity
sedimentation to examine the oligomeric states of a protein and
mutants thereof. Some mutants are trimers, and the molecular weight
estimates given by sedfit (either a c(M) distribution or a discrete
species model) are very reasonable. However, based on a crystal
structure, we expect the wild-type to be a 24-mer. Sedfit
consistently underestimates the molecular weight ( I get something
more akin to 21-mer).
I assume that there are at least to possibilites here:
1. The crystal structure is wrong, and the thing really is a 21-mer
in solution.
2. The vbar calculated by Sednterp is inaccurate- it is not
accounting for the fact that some of the volume of the 24-mer is not
taken up by protein, but by solvent. The vbar is therefore
significantly too low, with adverse effects on the MW calculation.
Does anyone know if there is a more accurate way to estimate vbar in
cases of large macromolecular assemblies? Can our crystal structure
help us out in any way?
BTW, yes, I know that sed. equilibrium might be the preferred
approach in this case, but instrument time is limited at the moment.
Thanks,
Chad
==================================
Chad A. Brautigam, Ph.D.
Research Scientist
The University of Texas
Southwestern Medical Center at Dallas
5323 Harry Hines Blvd.
Dallas, TX 75390
Office: (214) 645-6384
Fax: (214) 645-5383
_______________________________________________
RASMB mailing list
RASMB at rasmb.bbri.org http://rasmb.bbri.org/mailman/listinfo/rasmb
More information about the RASMB
mailing list