Fwd: Re: [RASMB] peak in sedimentation velocity scans at high protein conc.
Peter Schuck
pschuck at helix.nih.gov
Fri Oct 28 17:07:05 PDT 2005
Hi Everybody,
I agree with everything said, but since there seems to be a developing
general interest in this topic, I'd like to add one observation we made
repeatedly with the interference optics.
First - and this is certainly not new - one can easily visualize the
lensing effect of the boundary. I am attaching a powerpoint slide showing
the camera image from a run with several mg/ml thyroglobulin at 50000 rpm.
There's an obvious lack of light intensity in the boundary when using the
12 mm centerpiece, which is absent from the image of the 3 mm centerpiece
in the same run, as John mentioned. The lack of light intensity is what
causes the artificial peaks in the absorbance scans. [sorry, version
with the attachment bounced back awaiting approval ]
But the interesting point is that one can frequently still acquire fringe
shift data through the boundary - in fact even through the one shown in the
picture. They do not exhibit an artificial peak, but they have different
problems. At very high gradients, the scans seem to 'skip' fringes, i.e.
the algorithm to assign fringe shifts from neighboring points fails and
instead of an increase in fringe shift from one radial point to the next,
one may see even a slight decrease from one point to the next (I suspect
the XLI software simply can't deal with more than 0.5 fringes shift for
neighboring radial points). As Jack mentioned, we've seen this a lot also
in CsCl gradients, and at this occasion I've built into SEDFIT some limited
capability to repair this problem afterwards in software (with an algorithm
assuming that the boundary and its derivatives are monotonic). However,
there also seems to be a range of boundary steepness that is already
producing peaks in absorbance, but not yet exhibiting the fringe 'skipping'
in interference optics. Of course, the fringe profiles from these steep
gradients can still be questioned with regard to their faithful
representation of the concentration gradient, which I doubt.
However, the absence of the artificial peak at least allows us to load the
data and conveniently determine an empirical s-value from the midpoint.
Assuming that the distortion of the boundary shape is not modifying the
integral over the boundary in a time-dependent way, one might even be able
to get a fairly good precision on the weight-average s.
Best,
Peter
>X-SBRS: 5.5
>X-BrightmailFiltered: true
>X-Brightmail-Tracker: AAAAAA==
>Date: Fri, 28 Oct 2005 08:59:27 -0400
>From: Peter Schuck <pschuck at helix.nih.gov>
>User-Agent: Mozilla Thunderbird 1.0.6 (Windows/20050716)
>X-Accept-Language: en-us, en
>To: Norbert.Muecke at dkfz-heidelberg.de
>Subject: Re: [RASMB] peak in sedimentation velocity scans at high protein
> conc.
>X-Sanitizer: murus Advosys mail filter
>
>Hi Norbert,
>this is Wiener skewing from refraction of light within the boundary. If
>you use interference optics, you should not observe this. Also, you can
>probably minimize it by running slower rotor speed and not generating as
>steep gradients. Peter
>
>
>Norbert Muecke wrote:
>
>>Hi,
>>I have done some velocity runs at high protein conc. (up to 6g/l).
>>A new peak within the boundary could be observed.
>>I have attached the scans of three different concentrations.
>>Absorbance was measured at 284nm.
>>Any idea of how to explain that phenomena would be appreciated.
>>
>>Best regards
>>Norbert
>>
>>
>>Division Biophysics of Macromolecules, B040
>>German Cancer Research Center (DKFZ), TP3
>>Im Neuenheimer Feld 580
>>D-69120 Heidelberg, Germany
>>
>>
>>
>>------------------------------------------------------------------------
>>
>>Der folgende Teil dieser Nachricht enthält einen Anhang im sogenannten
>>Internet MIME Nachrichtenformat.
>>Wenn Sie Pegasus Mail oder ein beliebiges anderes MIME-kompatibles
>>Email-System verwenden, sollte Sie den Anhang mit Ihrem Email-System
>>speichern oder anzeigen können. Anderenfalls fragen Sie Ihren Administrator.
>>
>>The following section of this message contains a file attachment
>>prepared for transmission using the Internet MIME message format.
>>If you are using Pegasus Mail, or any another MIME-compliant system,
>>you should be able to save it or view it from within your mailer.
>>If you cannot, please ask your system administrator for assistance.
>>
>> ---- Datei Information/File information -----------
>> Datei/File: uebersichtscans.jpg
>> Datum/Date: 28 Oct 2005, 11:28
>> Größe/Size: 24596 bytes.
>> Typ/Type: JPEG-image
>>
>>
>>------------------------------------------------------------------------
>>
>>_______________________________________________
>>RASMB mailing list
>>RASMB at rasmb.bbri.org
>>http://rasmb.bbri.org/mailman/listinfo/rasmb
>>
More information about the RASMB