[RASMB] lamp intensity drop
Peter Schuck
pschuck at helix.nih.gov
Tue Jun 28 08:32:00 PDT 2005
This may not be related, but it reminds me on a problem we had on one
machine about a year ago, where mysteriously during a velocity run the
intensity at 255 nm (that's where we scanned) seemed to decrease and the
scans became very noisy. The first 20 - 30 scans were perfectly fine,
and then all scans were extremely noisy. When only a single scan was
started afterwards (cold) to test the intensity, it seemed fine, but
everytime during the velocity run it decreased again after a while. Our
service engineer was able to trace it down to what I seem to remember
was a space somehow created by the lamp housing or a cover, that was in
the light path but apparently didn't get evacuated and trapped O2
forming ozone after a while. (Sorry for the poor description, but I can
find out exactly what it was if somebody has the same problem.)
Peter
More information about the RASMB
mailing list