[RASMB] two technical questions
virgil schirf
schirfv at biochem.uthscsa.edu
Fri Jul 23 18:42:14 PDT 2010
Hi John,
The instrument should reach the set speed before any scans are
collected and should reach 40 krpm in less than 2 minutes when set to
400 rpm/sec (max) acceleration.
The monochromator motor is not precise and normally will skip increments
even if 1 nm is selected. I don't remember what the Beckman specs are
but +/- 4 seems a little high.
Best regards,
Virgil
jpsumida wrote:
> Hi RASMB,
> 1) I'm working with an XLI spinning an 8hole rotor fully loaded with 8
> cells running the instrument at 40krpm. It takes the instrument 20
> scans to reach top speed. This was determined by taking omega from
> omega-squared t, which I took from the second line of the header of
> each absorbance scan file. I'm not sure if this is the normal
> behavior of this instrument and it seems to me a fairly long period of
> time to wait for the instrument to come to speed. As I see a number of
> posts which note that software such as UltraScan correct data
> collected during the early part of a scan for acceleration effects,
> perhaps this is entirely normal behavior and not worth worrying about?
> 2) If one performs a wavelength scan with the instrument in absorbance
> mode, using a step size of 1nm increments, should one expect to
> observe in the data file absorbance data at 1nm increments? If one
> observes wavelength steps that vary +/- 4nm, could this be a problem?
> Thank you in advance for any comments you may wish to share, they are
> much appreciated.
> John Sumida
>
>
> _______________________________________________
> RASMB mailing list
> RASMB at rasmb.bbri.org
> http://rasmb.bbri.org/cgi-bin/mailman/listinfo/rasmb
>