[RASMB] Xl-I wavelength check question
Antje Voelkel
Antje.Voelkel at mpikg.mpg.de
Fri Feb 19 05:03:38 PST 2010
Dear Mark,
IF you consider to clean the slit assembly, please switch of the machine
before the PMT tube is uncovered. Our service told that it will be
damaged if the machine is running and costs ~5000 euro if I remember
correctly.
Best regards
Antje
________________________________
From: rasmb-bounces at rasmb.bbri.org [mailto:rasmb-bounces at rasmb.bbri.org]
On Behalf Of Mark Agacan
Sent: Wednesday, 17. February 2010 11:49
To: rasmb at server1.bbri.org
Subject: [RASMB] Xl-I wavelength check question
I have a question regarding the wavelength accuracy check for the XL-I:
Using a windowless cell (aluminium cell housing only with no centerpiece
or windows, or one of the empty holes in the rotor?) - how can I match
the weight of this cell with the counterbalance, even at the relatively
low speed of 3000 rpm?
The remainder of the procedure as outlined in the Troubleshooting and
Maintenance section of the XL-I manual seems straighforward enough, i'm
just a little unclear on exactly how to load the rotor for this
measurement.
Best Regards
Mark
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
Dr Mark Agacan, Scientific Officer for the Division of Biological
Chemistry and Drug Discovery,
Wellcome Trust Biocentre, College of Life Sciences, University of
Dundee, Dundee, DD1 5EH
Tel: +44 1382 386095 Fax: +44 1382 345764 Mobile: 07525 451 117
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
************************************************************
Please consider the environment. Do you really need to print this email?