[RASMB] Lamp intensity fluctuations
Titus M. Franzmann
tmfr at umich.edu
Tue May 4 07:11:29 PDT 2010
Hi,
I should have been a little bit more precise about my sample.
I loaded a cell with water (dH20) in the reference channel and a sample of
0.5 Abs in the sample channel. I recorded intensities instead of OD values.
Thereafter I plotted the lamp intensities from the reference channel (water)
as a function of the scan number using Ultrascan. The same graph can be
produced using Origin, plotting the raw intensities at any given radial
position, or the average of it, as a function of the scan number. Basically
the same profile.
Titus
-----Original Message-----
From: Leech, AP [mailto:apl3 at york.ac.uk]
Sent: Tuesday, May 04, 2010 9:52 AM
To: RASMB
Cc: Tom.Laue at unh.edu; Titus M. Franzmann
Subject: Re: [RASMB] Lamp intensity fluctuations
Hi Tom, Titus, all
I thought the "incident" readings come from the reference detector:
these can be seen using the menu Scan -> Request Data -> Incident
from the GUI, but they are only available from the last scan and
do not seem to be collected during a run. The incident data should
be flat across the scan, slightly noisy and slightly higher than the
maximum reference sector data (at least, that is how they are on our
XL/I). We once had a problem with the reference detector which resulted
in this data being a perfect flat line, solved by replacing the
monochromator assembly.
Best regards,
Andrew
Tom Laue wrote:
> Hi Titus, et al.-
> I am not sure what the intensity reading is you are presenting- is it
> the average value across the radial scan? Is it the reading at a
> particular radial position? In any case, it is worthwhile getting some
> idea of the complexity that is behind the XLA intensity readings.
> There are two intensity readings taken with each lamp flash- 1) 5 or 10%
> (can't recall which) of the lamp flash intensity itself is monitored
> using light bounced off of a semi-reflecting mirror and a photodiode
> sensor. This signal is used to correct for the inherent ~3% variation in
> flash intensity characteristic of the lamp. 2) The light intensity at
> the photomultiplier (PMT) after it has passed through the cell and
> slit/lens assembly. I do not believe the photodiode intensity readings
> are accessible to the outside world, so your readings are from the PMT.
[...]
--
Dr Andrew Leech * Laboratory Head
Technology Facility * Molecular Interactions Laboratory
Department of Biology (Area 15) * Tel : +44 (0)1904 328723
University of York * Fax : +44 (0)1904 328804
PO Box 373, York YO10 5YW * Email : apl3 at york.ac.uk
More information about the RASMB
mailing list