[RASMB] Lamp intensity fluctuations
Leech, AP
apl3 at york.ac.uk
Tue May 4 06:52:09 PDT 2010
Hi Tom, Titus, all
I thought the "incident" readings come from the reference detector:
these can be seen using the menu Scan -> Request Data -> Incident
from the GUI, but they are only available from the last scan and
do not seem to be collected during a run. The incident data should
be flat across the scan, slightly noisy and slightly higher than the
maximum reference sector data (at least, that is how they are on our
XL/I). We once had a problem with the reference detector which resulted
in this data being a perfect flat line, solved by replacing the
monochromator assembly.
Best regards,
Andrew
Tom Laue wrote:
> Hi Titus, et al.-
> I am not sure what the intensity reading is you are presenting- is it
> the average value across the radial scan? Is it the reading at a
> particular radial position? In any case, it is worthwhile getting some
> idea of the complexity that is behind the XLA intensity readings.
> There are two intensity readings taken with each lamp flash- 1) 5 or 10%
> (can't recall which) of the lamp flash intensity itself is monitored
> using light bounced off of a semi-reflecting mirror and a photodiode
> sensor. This signal is used to correct for the inherent ~3% variation in
> flash intensity characteristic of the lamp. 2) The light intensity at
> the photomultiplier (PMT) after it has passed through the cell and
> slit/lens assembly. I do not believe the photodiode intensity readings
> are accessible to the outside world, so your readings are from the PMT.
[...]
--