Antje Völkel Antje.Voelkel at uni-konstanz.de
Tue Dec 28 08:25:11 PST 2010


Dear Peter,

can you observe any oil in your maschine? Sometimes there is oil condensing if the AUC is cooled down.

An intensity scan of an empty rotor hole 1 might brin clearity if it condensed on the photomultiplier. You can also set the slit assembly to a more inner position if you set the radius in the HW parameters, IMPORTANT! switch of the maschine then!!! (to not damage the PMT tube) and remove the slit assembly and clean the PMT glas and may be the slit and the lenses of the assembly to.

Hope this helps!
Best regards
Antje


Am Dienstag, 28. Dezember 2010 16:46 CET, Peter Edward Prevelige Jr <prevelig at uab.edu> schrieb: 
 
> Hi All -
> 
> I may just be having post or pre holiday haze but I could use some help. I am trying to run a sed V on a sample for someone. 3, 2 sector cells. Protein OD at 280 is ~0.8 as reported and checked by me, wavelengtn is set for 280, scans at 3K don't make sense. Too high, too noisy, no meniscus, etc. Instrument was working fine last week. I've attached screen captures of the scan and the actual data file from cell 1.
> 
> Any suggestions on what to check to sort this out?
> 
> Thanks
> 
> Peter Prevelige
> 
> Peter
 
 
 
 