[RASMB] Protein charge
Tom Laue
Tom.Laue at unh.edu
Mon Dec 13 09:07:49 PST 2010
Hi all-
There is increasing interest in the accurate measurement of protein
charge since charge-charge and charge-dipole interactions contribute
significantly to solution behavior, particularly at high concentrations.
Some of you asked me for a quick review on measuring protein charge.
Three methods are available for measuring charge: capillary
electrophoresis, electrophoretic light scattering and membrane confined
electrophoresis. Each has its advantages and each has its limitations.
From my perspective, here are the pluses and minuses for each method.
_Capillary electrophoresis (CE)_
1. CE, pluses
1. Very small quantities required
2. True physical separation of charge isomers
1. Best method for detecting and quantifying charge isomers
3. Many (unused) instrument already are in labs
2. CE, minuses
1. Capillary wall interactions (e.g. electroendosmosis)
complicate data analysis
2. Solvent conditions often dictated by capillary wall
interactions
3. Requires hours to complete
4. Requires separate determination of the Stokes radius (by
Sed. velocity or dynamic light scattering)
_Membrane confined electrophoresis (MCE)_
1. MCE, pluses
1. First principle method, no standards required
1. Both mobility and steady state measurements are first
principle
2. Gold standard charge accuracy from steady state
3. Only method that can provide direct determination of
effective charge (steady state)
1. No frictional coefficient determination needed
4. Small volume (20 uL), at ~0.5 to 1.0 OD at 280
2. MCE, minuses
1. Steady state method is SLOW... complete set of measurements
require days to complete
2. Solvent interactions with membrane (di- and tri-carboxylic
acids) limit solvent choices
3. Steady-state method requires highly purified sample
_Electrophoretic light scattering (ELS)_
1. ELS, pluses
1. Fast, fast, fast! Requires ~5 minutes to measure the mobility
2. Wide range of solvent conditions may be used
3. Small volume (~200 - 500 uL) and low concentrations (< 1
mg/ml) required
4. Only method suitable for rapid survey work
5. Same instruments can provide Stokes radius via dynamic light
scattering
2. ELS, minuses
1. Scattering method, sensitive to aggregates
2. Accuracy and precision lower than MCE
1. Accuracy and limitations still being tested
3. Some early instruments were inaccurate due to technical problems
1. Manufacturers are working on fixes and testing
Commercial instruments are available for each of these methods:
CE- Beckman, Agilent, Applied Biosystems
MCE- Spin Analytical
ELS- Malvern, Wyatt, Brookhaven
I have had experience with some of these, but not all.
I'll be glad to answer any questions.
Best wishes,
Tom
--