Dear Sumit, dear all,
John Philo seems, addressed the right problemes of cell leaking !! 1) Since I use partly double-
sector centerpieces of the age over 30 years, they are some reasons that a cell may leak at several rotor speeds, even at zero rpm (waiting for vacuum before run-start.) In principle I put 2 housing-
plug-gaskets even at new centerpieces. For older centerpiece I put about 5 housing-plug-gaskets
so that the housing plug (small screw that we called 5 $ screw once they cost)) presses tight the (self cutted ) gasket and prevent liquids to escape (zero rpm) even in such low amount that you
don't see it on the vacuum indicator! ( sure the force should be limited, if not the cell swell so you will get difficulties install the cell in to the rotor hole). 2) In most cases I use Sapphire Windows,
since you can tourqe the cell a bit tighter then quarz ones. (Older centerpieces are not total flat,
so you flatten them a bit better). An additional reason is at highspeed SV speeds, that in case of on sector leaking that causes high pressure, the middle wall breaks and then usually the quarz window also breaks...3).I use in case of runs not over 30000 rpm !! usually SE runs
double-sector-centerpiece - gaskets, (it is quite hard to install them and they are very expensive)
those used in alum/metal DS'tors, but it is possible also if you find the trick to do it, so you prevent leakages ( as also John menthioned) between the sectors with all his consequences.
yours ariel
ariel.lustig at bluewin.ch