> Dear All,
>
> We're having a serious problem with our XL-I. The UV filter that is
> positioned over the condenser lens is fogging with oil every time it
> is used. It is the underside of the filter that gets fogged.
> Fogging, of course, completely negates our ability to use the
> interference optics. A Beckman service rep recently changed the
> condenser lens and also replaced our vacuum pump. Does anyone have
> any notion of why this fogging is occurring and what the fix might
> be? Our service rep is trying hard to get this fixed, but he is new
> to the analytical centrifuges. Any help the community can offer
> would be greatly appreciated.
>
> Chad
Hi Chad,
oil condensation seems to be a common problem with the XLA/I. In my experience
this problem is minimized when running at lower temperature, and aggravated
when running at temps higher than 25C. I think the only surefire solution
is to replace the pump with a turbomolecular pump. You may want to check
with Walter Stafford, I believe he has made such a retrofit to his machine.
Regards, -Borries