There is a missing vbar term, k(P) should be = 2BM - vbar, an error in Tanford's otherwise powerful book. Paley Johnson and myself looked at this in 1984 in a paper published in Biochemical Journal:
http://www.nottingham.ac.uk/ncmh/harding_pdfs/Paper28.PDF
It is not significant for highly concentration dependent systems such as polysaccharides, but for globular proteins it can be,
Steve
 

________________________________

From: rasmb-bounces at rasmb.bbri.org on behalf of Arthur Rowe
Sent: Mon 08/09/2008 12:28
To: Christine Ebel; 'Beld, Joris'; RASMB at rasmb.bbri.org
Subject: Re: RE : [RASMB] upper concentration limit AUC



	Hi Christine (and everyone)
	
	I think this discussion is already in the RASMB Archives.
	
	To summarise very briefly, k(D) is indeed lower than k(s), but not hugely so. For simple, spherical particles, defining all coefficients in volume fraction terms and at limiting infinite dilution, we can write
	
	k(D) = k(P) - k(s) where k(P), the concentration dependence of the chemical potential which is the force driving the flux in translational diffusion, and is given for s single component by (2BM + higher terms) where B is the 2nd virial coefficient. For spheres, 2BM = 8 ml/g; k(s) = 4* ml/g:  so k(D) = 4 ml/g. I am afraid that neglecting k(D) is not a good approximation. 
	
	Kind regards - will be seeing you (Christine) and lots of others in Newcastle this week. 
	
	Arthur
	
	*see "The sedimentation rate of disordered suspensions"  Brady, John F.; Durlofsky, Louis J.
	Physics of Fluids 1988 31 717-727 for the last word from the fluid mechanics people. You can also look up my Chapter in the 1992 Book (the 'black book') ed s Harding, Rowe & Horton) for details of my own derivation. Which happens to be in total numerical agreement with the Brady/Durlovsky treatment, right p to 64% volume fraction . . . . . 
	Note: the value given above for k(s) = 4 ml/g is for DYNAMIC k(s). The usual measured value which gets reported is the KINEMATIC k(s), which for spheres has the value 5.0 ml/g. You can get either value out of theory, depending how you play it. Brady & Durlovsky report the kinematic value. All of which was first noted by Burgers in 1939/40, and has been much ignored ever since! As no density 'correction' is called for in the estimation of k(D) or of k(P), it seems reasonable to use the dynamic k(s) in the equation above.
	
	


	Dear Joris, dear all, 
	
	This reference may be perhaps useful. 
	
	
	
	Solovyova A., Schuck P., Costenaro L., Ebel C, 
	
	Non-ideality by sedimentation velocity of halophilic malate dehydrogenase in complex solvents, 
	
	(2001) Biophys. J, 81 1868-1880. 
	
	
	
	In this work, we analyzed sedimentation velocity profiles considering hydrodynamic and thermodynamic non ideality. (i.e. concentration dependency of s and D) in the case of an homogeneous  solution of our protein of interest. The modified Lamm equation was implemented in a model of analysis in Sedfit (note that, unless the programme was recently modified, the kS and kD are expressed in signal unit in sedfit) 
	
	From my experience, the sedimentation velocity profiles are essentially modified by the concentration dependency of the sedimentation coefficient. Thus the concentration dependency of D can be neglected in a first approximation. Also from a theoretical point of view, kD is much lower that ks. 
	
	All the best 
	
	Christine 
	
	
	Christine EBEL 
	
	Institut de Biologie Structurale CEA-CNRS-UJF 
	
	41 rue Jules Horowitz, F-38027 Grenoble France 
	
	Tel (33) (0) 4 38 78 95 70; Fax (33) (0) 4 38 78 54 94 
	
	christine.ebel at ibs.fr 
	
	http://www.ibs.fr/content/ibs_eng/presentation/lab/lbm/ebel.htm 
	-----Message d'origine-----
	De : rasmb-bounces at rasmb.bbri.org [mailto:rasmb-bounces at rasmb.bbri.org] De la part de Beld, Joris
	Envoyé : lundi 1 septembre 2008 15:24
	À : RASMB at rasmb.bbri.org
	Objet : [RASMB] upper concentration limit AUC 
	
	
	
	Dear all, 
	
	
	
	A colleague asked me whether analytical ultracentrifugation has an upper limit with regard to the concentration of the protein. They want to measure the protein at the same concentration as the NMR experiments (> 1 mM). I am not entirely sure but I thought this should be no problem. One could easily measure off-peak at another wavelength than 230nm, e.g. 235nm or 280nm, right?! Or does one run into non-ideality phenomena when doing sedimentation equilibrium at these high protein concentrations?! 
	
	Thanks a lot in advance for any feedback. 
	
	Best wishes, 
	
	
	
	Joris Beld 
	
	
	
	Hilvert Group 
	
	ETH Zürich 
	
	Switzerland 
	
	_______________________________________________
	RASMB mailing list
	RASMB at rasmb.bbri.org
	http://rasmb.bbri.org/mailman/listinfo/rasmb
	
	





This message has been checked for viruses but the contents of an attachment may still contain software viruses, which could damage your computer system: you are advised to perform your own checks. Email communications with the University of Nottingham may be monitored as permitted by UK legislation. 


This message has been checked for viruses but the contents of an attachment
may still contain software viruses, which could damage your computer system:
you are advised to perform your own checks. Email communications with the
University of Nottingham may be monitored as permitted by UK legislation.

-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20080915/042fb71a/attachment.htm>


More information about the RASMB mailing list