Dear Christine, dear Colleagues,
it may be, that in your case you have to work in a D2O - buffer, where the sed.velocity me be reduced
due to higher solvent density but the floataion may increase. Sure one have to try different densities
where you may catch both fractions in the same run or to catch one out of the field.
Apro-pos the different densities may help you
for the determination of the density dependent fractions (seeC H Huang and J Charlton (JBC 1971 vol 216 8 April ) If you speak about purification I understand that you intend to separate preparativly
some fractions. At this point I like to mention that they are Vertical rotors where the tubes have approximately the same diameter as an UC cell.so , if you have the run-condition , speed and time
of the analytical run, you can copy this condition for the vertical rotor. When the rotor stands still the fractions (bands) became horizontal and easy to suck them out.
I agree 100 % with Kristian Schillig about the schlieren-peak inside the abs.phenomena, but this knowledge will not help to separate or determe this two different moving boundariesbut to recgnize them.
good luck Christine.....ariel