While there is no practical upper concentration limit to the
interference optics, there are very real limits to the concentration
gradients it can follow. The horizontal pixel spacing of the camera and
the "splicing" routine used to convert fractional fringe displacement to
the total fringe displacement limit the concentration gradient to values
less than 0.5 fringes/10 um (500 fringes/cm). While this seems like a
lot, it is in the range of concentration gradients where focusing
affects the accuracy of the concentration determinations (e.g. Weiner
skewing may be significant).
My experience is that with 12 mm centerpieces, the system has difficulty
tracing gradients in solutions with concentrations > ~20 mg/ml
(depending on how much or little diffusional broadening there is).
If you are less concerned about the accuracy of the concentration
readings, you may increase the concentration range by using thinner
centerpieces. Be aware, though, that the focus of the XLI assumes 12 mm
centerpieces are being used. You will still get good fringes, and the
concentrations will be reasonably close to the correct values so long as
the gradients are not very steep. You can get an idea of how much error
to expect from David Yphantis's 1964 paper:
D. A. Yphantis. Equilibrium Ultracentrifugation of Dilute Solutions.
Biochemistry 3:297-317, 1964.
Best wishes,
Tom
christine.wandrey at epfl.ch wrote:
> Dear all,
>
> I started analyzing micelle/particle polymer systems at relative high
> concentrations by interference optics. The aggregation is concentration
> dependent and needs certain minimum concentration.
> During the first screening experiments I realized strong fringe shifts 20-40
> fringes.
> I know that the Rayleigh optics has some limitations but I could not find
> quantitative numbers.
> Thank you for help.
> Best,
> Christine
>
> _______________________________________________
> RASMB mailing list
> RASMB at rasmb.bbri.org
> http://rasmb.bbri.org/mailman/listinfo/rasmb
>
>
--
Department of Biochemistry and Molecular Biology
University of New Hampshire
Durham, NH 03824-3544
Phone: 603-862-2459
FAX: 603-862-0031
E-mail: Tom.Laue at unh.edu
www.bitc.unh.edu
www.camis.unh.edu
More information about the RASMB
mailing list