DTT as well as other reductants absorb in the UV. In addition, reductants
tend to change their extinction coefficient between reduced and oxidized
states. Take a look at this reductant table:
http://www.cauma.uthscsa.edu/ (click on "Resources", then "Reductants")
The more your buffer absorbs, the less dynamic range in the absorbance
is available to your protein. The XLA's spectrophotometer is only good
between about 0-1 OD, so if you use up half of this for your buffer
absorbance (essentially, raise the baseline absorbance) you don't have
much left for sample absorbance.
There are some alternatives: If you have to have reductant present
you can measure pretty much without problem at 280 nm if you replace
DTT with TCEP.
Good luck,
-Borries
>