The spacers for the 3 mm centerpieces place the centerpiece at the
mid-point of the cell, which is out of focus for the interference
system, which is focused at the 2/3 plane of a 12-mm centerpiece, with
sapphire windows and a solvent refractive index of water. Mis-focusing
will lead to mis-registration of the true radial position with the
apparent radial position in regions of the image with steep gradients in
the concentration (1). For the 3 mm centerpieces, the focal plane is
below the sample 2/3 plane, causing steeper fringes than might be
expected... i.e. analysis will reveal what appears to be a weak
association or a heterogeneous system. This problem would also affect
the absorbance system, but the absorbance optics cannot trace a steep
enough gradient.
So long as you keep gradients low enough (as I recall, about 50 fringes
displacement per mm), your data will be OK.
Best wishes,
Tom
(1) D. A. Yphantis. Equilibrium Ultracentrifugation of Dilute
Solutions. Biochemistry 3:297-317, 1964.
Borries Demeler wrote:
> Question about 3 mm centerpieces - the spacers that come with them seem
> to take it out of the focal plane for the interference system, at least
> we cannot get 3 mm CPs to work in interference mode. Are there spacers
> that will correct this issue, or is there something else we should try?
>
> thanks, -Borries
> _______________________________________________
> RASMB mailing list
> RASMB at rasmb.bbri.org
> http://rasmb.bbri.org/mailman/listinfo/rasmb
>
>
--
Department of Biochemistry and Molecular Biology
University of New Hampshire
Rudman Hall 379
46 College Road
Durham, NH 03824-3544
Phone: 603-862-2459
FAX: 603-862-0031
www.camis.unh.edu
www.bitc.unh.edu
More information about the RASMB
mailing list