[RASMB] ellipsoids of revolution
Tom Laue
Tom.Laue at unh.edu
Fri Sep 29 08:32:33 PDT 2006
Hi-
Only one hydrodynamic parameter may be extracted rigorously from the
sedimentation coefficient extrapolated to zero solute concentration, and
that is the frictional coefficient, f, from which the Stokes radius, Rs,
may be calculated:
f = 6 pi eta Rs, where pi = 3.14, eta is the solvent viscosity and Rs is
the Stokes radius.
There are several things to note about this-
1. f = 6 pi eta Rs for a solute that exhibits "stick" conditions with
the solvent (i.e. there are favorable solvent/solute surface
interactions). For "slip" conditions, f = 4 pi eta Rs. Where it
has been tested, biological molecules exhibit stick behavior.
Given what is known about the solvent interactions of
biomolecules, I am comfortable that there is no need to worry
about using 6. FYI, for rotary diffusion, the 4 pi term is the one
to use.
2. The solvent viscosity is used because the calculation of Rs from f
becomes extremely complicated at higher concentrations. By using
very dilute solutions, or extrapolating s to c=0, there is no
ambiguity about what viscosity should be used. At high
concentrations, the bulk viscosity is not an appropriate term for
interpreting the frictional coefficient. To learn more, I
recommend Fujita's monograph.
3. The Stokes radius comes from the solution of the Navier equation,
using Stokes approximation (an incompressible solvent). Solving
the Navier-Stokes equation involves setting a boundary that
encloses one volume of solution from another, then applying the
conservation principles for mass, energy, linear momentum and
rotary momentum. There is little that is instructive in going
through the math involved in the solution. One important point is
that conservation of energy includes conservation of charge, and
that conservation of momentum includes ion-ion momentum transfer
through space.
4. The Stokes radius has been interpreted further, but only by
imposing a model. Perrin determined solutions to Stokes equations
from which the axial ratios for _solid, uncharged_ molecules could
be calculated from the Stokes radius.
5. Calculation of axial ratios from the Stokes radius inevitably
results in ratios that are larger than would be expected (or that
are known from other data). A rational explanation for the
discrepancy is that there is a layer of solvent carried with the
molecule, and that this solvent layer adds to the molecular size.
Hence, the inclusion of the "solvation" layer, usually described
in terms of the number of grams of water bound per gram of
protein. But this approximation needs to be viewed with care since
it is not a rigorous interpretation of Rs as it is not developed
from the Navier-Stokes equation. For starters, there is only one
parameter accessible from s, and now we are trying to get two- the
axial ratio and the extent of solvation. One of these two terms
must be set.
6. There is some confusion about the difference between solvation and
hydration. Sednterp estimates the _hydration_ of a protein using
data from Kuntz and Kauffman obtained by proton NMR. Steve Perkins
has shown that the hydration of pretty much every protein is about
0.3 g/g. However, it is the _solvation_ of the protein that
matters in Rs (and even that is a bit of a kludge), and the
solvation includes the through-space momentum transfer due to
charge coupling. Hence, if you use the hydration values calculated
by Sednterp, or if you use Steve's value for hydration, you will
almost inevitably end up with axial ratio estimates that are large
compared to estimates from other methods.
The long and short of it is that Rs is a valid and valuable quantity
describing both the asymmetry of a protein and its interactions with the
solvent. Comparisons of Rs values for a protein acquired in different
solvents, or between molecules is valid. One should be careful, though,
about interpreting Rs solely in terms of the shape of a molecule.
Best wishes,
Tom
mitrana at mail.utexas.edu wrote:
> Hi all
>
> I have a very basic question. How does one get from determining the a/b ratio to
> the gross molecular dimensions of the protein molecule. The molecular weight
> should place a limit but I am a bit fuzzy as to how exactly this is carried
> out.
>
> Mitra
>
--
Department of Biochemistry and Molecular Biology
University of New Hampshire
Rudman Hall 379
46 College Road
Durham, NH 03824-3544
Phone: 603-862-2459
FAX: 603-862-0031
www.camis.unh.edu
www.bitc.unh.edu
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20060929/2c9b343c/attachment.htm>
More information about the RASMB
mailing list