Hi John,
Thank you for your comments.
I think it is OK to say that the data is absorbance and that it looks like
a picture from a Beckman training manual. I was a little surprised they
were interested enough in the truth to let me describe the problem I was
having on RASMB in this general way to get community input, but I don't
want to push things by divulging any details: I actually do not know
myself what the buffer is. I mention non-ideality because the darn scans
just don't fit as a whole but each one looks fine by itself.
I might be more interested in getting a good fit than the people who own
the data: but it really bugs me that something that looks so simple
doesn't fit the models I tried in Sedfit and Sedanal.
Thanks
David Hayes
>David,
>
>It is hard to comment without knowing more about the experiment or seeing
>the data. Can you at least tell us whether you are talking about
>absorbance or interference data?
>
>You mention that you have considered non-ideality, which suggests that
>this could be an experiment at high protein concentration. Is that
>correct? If so, then part of the problem might be optical distortions due
>to refraction by the boundary---these are always worst early in the run,
>and go away as diffusion makes the gradient less steep. That problem can
>be significantly reduced by using 3 mm centerpieces.
>
>John
>-----Original Message-----
>From: rasmb-bounces at rasmb.bbri.org [mailto:rasmb-bounces at rasmb.bbri.org]
>On Behalf Of David Hayes
>Sent: Friday, March 24, 2006 12:41 PM
>To: RASMB
>Subject: [RASMB] Sedimentation Velocity Experiments that are hard to fit
>
>Hi to all experienced Ultracentrifuge Scientists,
>
>I have some data (repeated experiments) from sample A buffer 1 (which may
>or may not have been produced on the planet earth -- that is all I can say
>about it)
>
>There are two puzzling aspects to this data.
>First, I am not getting good repeatability in the amount of aggregate.
>Second, by eye, the raw data looks like a single species (which it should
>be) with a small amount of higher weight aggregate (not surprising), but
>the entire data set never fits well no matter what I try.
>Thinking that the repeatability problem might be an artifact of my
>inexperience using Sedfit I tried many different strategies: including
>floating or not floating just about every parameter and also using the
>nifty "experimental initial distribution" option of Sedfit.
> Fitting to the whole data set, I got a fit with a single large peak and
> two very small aggregate bumps which all together integrate to about 4%
> of the total. The rmsd was a bit large at 0.01 and the residuals were
> systematic. Early scans were fit very poorly and later scans fit
> somewhat poorly with scans in the middle fit pretty well. Later using
> Sedfit with the single species model and only the later half of the scans
> and floating a lot of things and turning off the hat function in Sedfit
> brought me down to a comparable fit with a rmsd of .006, but the
> residuals were still systematic and the molecular weight was off more
> than what I expected for having 4% aggregate.
>Then I did something that Peter Shuck thinks is a bit strange, but I think
>it helped me understand the data a bit more. I fit pairs of scans with
>c(s) taking scans at different times paired with the last
>scan. Essentially then, I was fitting each scan individually: using the
>last scan in ever pair was needed to help Sedfit know the baseline. Peter
>reminded me that by fitting each scan, I lose the ability to distinguish
>heterogeneity from diffusion. But in this case, all the evidence points
>to a single species, so I did not think heterogeneity (beyond the measured
>4% aggregation) was significant. The peak S value stays constant all the
>way down the cell. Each individual scan fit this way fit very well
>without systematic residuals and a rmsd of about .005. What changed was
>that early scans fit to an f/f0 of 1.06 while by the later scans f/f0 fit
>to 1.5. This means that the early scans are broader than they should be
>for a reasonable f/f0 and that the later scans seem to be narrower and
>diffusing less.
>I repeated this with Sedanal dcdt based curve fitting (here I took scans
>by twos going down the cell and fit to a single species ignoring the 4%
>aggregate) and found a similar trend: the fitted molecular weight went
>from 95,000 to 225,000 and then settled down to about 160,000. If there
>were heterogeneity, the scans would fit the opposite way to a one species
>model, the f/f0 would go down, or the weight would go up later in the run
>as the species sedimented apart.
>
>I had some interference data of Tropomyosin that I subjected to a similar
>type of analysis with Sedanal, fitting the whole data set and then fitting
>scans by sets of 10 for the different times in the experiment (because of
>TI, RI noise fitting, it is not possible to fit small sets of scans with
>Sedfit using interference data; however, a Sedfit fit to the whole Tm data
>set showed no systematic variation of the residuals dependent on scan
>number). There was no scan number dependent trend in molecular weight
>with this molecule, though the variability in fitted molecular weight was
>surprising (weights from different sets of 10 varied from 60,000 to 80,000
>randomly).
>
>Seeing this I tried a Sedfit single species and Sedanal single species
>fits with non-ideality parameters, but floating these parameters made no
>substantial difference.
>
>I don't know if anyone has run into this type of problem, but I am out of
>ideas what to try next. I am being a bit stubborn wanting a good fit to
>the data: the S value of the main peak is reproducible and the data
>contains quite a bit of information, but I can't figure out exactly what
>is going on:
>
>Is this just non-ideality that I am not fitting properly or maybe that is
>not modeled well?
>Is the culprit convection? Thoughts on convection in the next message.
>
>Cheers
>
>
>Dr. David B Hayes
>Boston Biomedical Research Institute
>64 Grove St.
>Watertown, MA 02472
>617-658-7738
>
>Boston Biomedical Research Institute... Today's Research for Tomorrow's
>Health.
>Please visit us at www.bbri.org
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://list.rasmb.org/pipermail/rasmb-rasmb.org/attachments/20060324/1dd5423d/attachment.htm>
More information about the RASMB
mailing list