[RASMB] One pure sample, three wavelengths, and differing apparent elution volumes !
Richard Kingston
rl.kingston at auckland.ac.nz
Fri Feb 3 17:22:26 PST 2006
Recently, when running molecular weight standards over a size
exclusion column I made a puzzling observation.
I routinely monitor column outflow at three wavelengths, since our
chromatography system allows it (I'm using an Amersham Biosciences
Äkta). This is to ensure that an undistorted elution profile is
obtained at at least one wavelength (For concentrated, and highly
absorbing samples, the detector response may become non-linear). So
typically I would monitor the signal at 280 nm (the peak due to Trp/
Tyr/Cys) and at then on the shoulder of this peak (293 and 300nm) for
reduced sensitivity. Or when increased sensitivity is required, I
often monitor at 220 or 230 nm, on the far shoulder of the peptide
bond absorbance peak.
In the past, when using the instrument in this way, traces have
always been identical (up to some scaling factor) when the absorbance
is < 1.0.
But last week, when running standards over a Superdex 75 column, one
of them (Ribonuclease A) gave different apparent elution volumes at
different wavelengths (280, 293 and 300nm). The behavior was
reproducible, and the shifts in the peak position were small yet
significant (far beyond the inherent uncertainty in the peak
position). Apparent elution volume at 280 nm > 293 nm > 300 nm. This
also corresponds to the order in which the wavelengths are cycled by
the monochomator.
Yet the remaining standards (Chymotrypsinogen A, Ovalbumin, Albumin,
Blue Dextran 2000, L-Tryptophan) , run before and afterwards, in an
identical fashion, did not exhibit this behavior. In these cases the
traces at different wavelengths were all equivalent, accounting for
the differing absorbance by the sample. These standards are both
smaller and larger than Ribonuclease A.
So the behavior is both puzzling and concerning (what wavelength
should I 'believe' in the case of Ribonuclease A?) and I'm left
scratching for an explanation. I'm hoping somebody on the list might
suggest something.
The UV Monitor on the Akta has a 1 cm path length continuous flow
cell. Light is provided by a xenon flash lamp, monochromated with a
grating, and there is a beam splitter which sends 50% pf the light
through the cell, and the remainder through a reference fiber. Up to
three wavelengths can be recorded 'simultaneously' (in practice the
wavelength is rapidly switched, with the reported switch time < 500 ms).
Thanks for any suggestions
Rich.
Richard Kingston, PhD
School of Biological Sciences
The University of Auckland
Auckland
New Zealand.
rl.kingston at auckland.ac.nz