[RASMB] torque strength
Walter Stafford
stafford at bbri.org
Thu Nov 17 08:04:06 PST 2005
Dear RASMBers
With regard to the UV absorbance at short wavelengths, I would like
to add that we take spectra and catalogue them for all our sapphire
windows. The absorbance at short wavelength varies quite a bit from
window to window. With a careful choice of windows, one can work down
to 220 nm or lower. Some windows however are not much good below
about 250 nm. If I remember correctly, the original specification
from Beckman through Meller Optics in Rhode Island, was that the
sapphire windows be sufficiently transparent in the UV to be used
with the UV scanning optics on the Model-E.
Walter
p.s. I just got off the phone with Tom at Mellor Optics and he will
email me the specs on the both the standard windows and the UV clear
windows. I will share these with you when I get them.
At 15:20 +0000 11/17/05, pjgb at mrc-lmb.cam.ac.uk wrote:
>Dear RASMBers,
>
>Ariel's advice is fine as long as one does not wish to use short UV
>wavelengths, where the absorption of the sapphire windows in too
>high.
>Possibly replacing distorted epon centrepieces is simpler and cheaper.
>
>Jo
>
>--On Thursday, November 17, 2005 4:05 pm +0100 Ariel Lustig
><ariel.lustig at bluewin.ch> wrote:
>
>>
>>Hi Karen and colleagues,
>>I can tell you only from my own experience how I treat this problem.
>>First of all I use only sapphire windows, since at the end they are
>>cheaper then many broken quartz windows. Why broken ? my cell-stuff is
>>very old so Epoxy centerpieces (alum-filled and charcoal-filled) swell
>>slowly and became a bit convext on both flat side , so I even install
>>(not too easy) in case of the bad once DS-window gaskets between the
>>window and the centerpiece. as it is obligatory for metallic
>>centerpieces!!
>>Sure that in case of metllic centerpiece the torque strength is more
>>crucial since lack of elasticity you break easier the window just
>>during closing the cell (not the centerpiece as you wrote ,Karen ! )
>>so 130 torque -units should be enough. A probleme may be the
>>temperature, I torque the cell at the (coldest) temperature the run is
>>planed , independend if I fill the cell at room temp.( if the sample
>>allows it). The idea is to prevent shrinking later of the
>>centerpiece....and leaking. A bit tolerance for higher temp. is still
>>given with 130 t. The Epoxy once "suffer" more tolerance, so 120 -140
>>is OK.
>>With time usually one gets the feeling witout looking on the torque dial
>>especially using Epoxy once , according the flexibility.
>> I don't believe that the torque instrument is a so precise instrument
>>and unic by all users !!! It may happen (I believe at most of you) that
>>the screw-ring -gasket do not lie properly on top of the upper window-
>>holder (catched in the winding or an alien element is between) breaks ,
>>and as response you may read a high torque dial but in realty the
>>centerpiece is not pressed to the windows.
>>Just returning to centerpiece break during a run , usually happens
>>with quartz widows. First you have a leak in one sector (may be
>>evaporation true the filling screw), the filling hight liquid remain
>>much higher as in the other sector, then the midle strap breaks and also
>>mostly the quartz window. (Saphire even at max.speed never break.)
>>yours...ariel
>>!
>>
>
>
>
>P.J.G. Butler,
>MRC Laboratory of Molecular Biology,
>Hills Road, Cambridge, CB2 2QH, UK.
>Tel. +44 (0)1223 402296