[RASMB] radial shift
Tom Laue
Tom.Laue at unh.edu
Wed Jun 29 17:27:00 PDT 2005
Hi Joris,
It sounds very much as though you have a problem with the radial
positioner on your instrument. The servo mechanism can become gummed up
after while, and the pinion gear can loosen. From your description, it
sounds like the former problem is present. I would have the Beckman
engineer out to check it out.
You might try analyzing the data. However, unless you are interested
only in a simple question (e.g. is this material monomer or dimer), I
would not spend a great deal of effort in the analysis.
Best wishes and sorry!
Tom
Beld, Joris wrote:
>----------------------------------------------------------------------------------
>The older archived RASMB emails can be found at:
>http://rasmb-email.bbri.org/rasmb_archives
>and current archives at
>http://rasmb-email.bbri.org/pipermail/rasmb/
>Search All the Archives at:
>http://rasmb-email.bbri.org/rasmb_search.html
>----------------------------------------------------------------------------------
>
>Dear all,
>
>Lately we have encountered a strange problem with our XL-A. For a SE run we normally scan three times (with ten averages) per cell per wavelength. In principle these three scans are not further than 10-20min apart. Still, we can see a relatively big radial shift (0.2-0.4 mm). The scans are superimposable if shifted over the x-axis. I was doing two consecutive runs (with different proteins at different speeds, from 4k to 30k) with scans both at 235 and 280nm and this strange behavior is more pronounced at 235nm than at 280nm. Of course, one could think of samples that have not reached equilibrium yet but we estimated the time to equilibrium with ultrascan and let the system equilibrate varying from three days to 20h. So, I think the samples are at equilibrium, and moreover, within 10-20min one should not see such big radial shifts.
>
>We have actually two questions:
>- do you think we can still use this data? (in principle fitting should not be influenced by radial shifts, or am I making a mistake in that assumption?!)
>- does anyone has some experience with this problem and/or knows a solution (before I call Beckman)?
>
>Thanks a lot in advance.
>
>To comment quickly on lamp cleaning. We clean the lamp monthly (with toothpaste) and so far that seems to be sufficient.
>
>Kind regards,
>
>Joris Beld
>
>Swiss Federal Institute of Technology
>Zürich, Switzerland
>_______________________________________________
>RASMB mailing list
>RASMB at rasmb-email.bbri.org
>http://rasmb-email.bbri.org/mailman/listinfo/rasmb
>
>
>
--
Department of Biochemistry and Molecular Biology
University of New Hampshire
Rudman Hall 379
46 College Road
Durham, NH 03824-3544
Phone: 603-862-2459
FAX: 603-862-0031
www.camis.unh.edu
www.bitc.unh.edu
More information about the RASMB
mailing list