[RASMB] Cell Leak
Virgil Schirf
schirfv at biochem.uthscsa.edu
Wed Aug 24 14:48:01 PDT 2005
Allen,
We've seen the rib break but not bend as you describe. The centerpiece
is probably now corrupted especially for speeds that far above its
maximum rating. I would worry that if any resulting surface
irregularities developed during bending, eddying effects may occur even
if the centerpiece remained intact. Empirically, you might be able to
check the integrity of the centerpiece by running something like BSA and
critically examining the data.
You might want to consider purchasing centerpieces that are rated for
use up to 60krpm such as titanium, aluminum, or the SedVel60k graphite
filled epoxy centerpieces.
Best regards,
Virgil
Allen H Pekar wrote:
>
> RASMB members: We had a cell leak today where all the liquid leaked
> from one sector but little leaked from the other sector. As a result
> of spinning at 60,000rpm, the plastic between the two sectors was
> bent. When the cell was taken apart, the plastic returned to its
> original shape. Has anyone had this happen and if so, is it safe to
> use the centerpiece? We mostly do velocity runs at 60,000rpm.
> Thanks, Allen Pekar (Eli Lilly and Company)
--