the disulfide chromophore has an extinction coeifficient of around 110 at
280. Now if one has 200 mM DDT, and 10% is oxidized, then one has a
background buffer absorbance of 2.2 at 280.
My counsel is to always run a bench top absorbance scan of both buffer and
sample if it is a sample given to you. I think all of us have horror stories
of hysterical samples.
As for detergents, I've seen problems with sed eq data at higher speeds when
the micelles redistributed enough to throw the light beam off center towards
the bottom.
There is no substitute for looking at the intensity of the reference sector
for equilibrium absrobance scans. This will tell you the story. I take
credit for this being in the Beckman acquisition software.
best regards
Les Holladay