Dear Ariel,
Your SEGAL program arouses my curiosity. I find that it is very user
friendly. Beckman's software is OK to use, but editing data to achieve a
good variance is a pain in the neck. I assume your correlation value is
related to the variance. The idea of simulating a good correlation value
simultaneously with a dynamic graph before your eyes is very cool and heaven
sent, I like it. I hope that you can expand your program to include
analyzing six-channel equilibrium cell, which is often used for denaturation
studies, all in one single page of data. I can analyze two cells now, print
out data and then do another cell to print another page data. I think you
can create part 3 for the third cell, even though this will make all graphs
smaller but the trade off is great, you can include all three-cell (or
so-call six-channel) data in one single page. It will save space and time
and make life simple. Using Beckman's software, in this case, is much more
tedious. You need to select and save three subsets. You then have to go to
equilibrium program to analyze each saved subset, which may or may not get
the best variance value.
I was wondering if SEGAL program could do multiple fitting (global fitting).
Also can one go head and calculate association constant with this more
accurate apparent molecular weight obtained by SEGAL. If not, do you have
any plan to add that capability in the future?
Thank you for your contribution to create a much needed user-friendly
program.
--------------------------------------------------------------------------
Christopher Chin
Manager, XLA-Analytical Ultracentrifugation facility
Sealy Center for Structural Biology
HBC&G, 5.134 MRB.UTMB, Galveston, TX 77555-1055
cchin at utmb.edu, 409-772-1693, efax 708-585-1920
---------------------------------------------------------------------------
-----Original Message-----
From: Lustig [mailto:ariel.lustig at unibas.ch]
Sent: Wednesday, June 26, 2002 7:49 AM
To: rasmb at rasmb.bbri.org
Subject: [RASMB] 2 subjects, overspeed, " SEGAL"
----------------------------------------------------------------------------
------
The older archived RASMB emails can be found at:
http://rasmb-email.bbri.org/rasmb_archives
and current archives at
http://rasmb-email.bbri.org/pipermail/rasmb/
Search All the Archives at:
http://rasmb-email.bbri.org/rasmb_search.html
----------------------------------------------------------------------------
------
Dear RASMB colleagues ,
Dear Andrew,
Subject 1
in the "Manual of Methodes" (C H Chervenka) for the AUC Spinco -Beckman
Division-Palo Alto
(for the model "E") is the Overspeed -technic discussed, page 44 and
ref.75:
Hexner P E, Radford L E ,Beams J W "Achievement of SE "
Proc. Natl. Acad. Sci. 47, 1848 (1961)
As you ask for a practical procedure=rule of thumb, I like first of all to
emphasize
that the time to reach equilibrium depence on three parameters, 1)The
filling height
of the cell, 2) the diffusion constant of the solute=substance . 3) the
viscosity of the solution.
The optimal over-speeded rotor speed and time is, that still at least a
1/3 to 2/3 of the plato,
(the region that in case of absorption shows approximately the
C(0)=initial concentration
you filled and tested it at very low speed , just at the run -start or
compaire it with the wavelength
scan at the same nM value.(Be a bit carefully about the WL baseline.)
What are the dangerous points on have to be careful : at a too relative
high speed the sample
may be dragged too far to the cell bottom , so you loose the time
-spare advantage,
but worth is if the particles
or molecules du not reversibly back-diffuse (means insoluble aggregates)
due to high
concentration near the bottom.
On principle we have to look several times during the overspeed -
procedure on the partly
depleted part near the meniscus and the unreached bottom cell part.
Subject 2.
I would like to remind you that at the 23 of may 2002 I sent to all of
you our =(Biozentrum)
SEGAL computer that can be loaded down from the net
http://www.biozentrum.unibas.ch/personal/jseelig/AUC/index.html
I would like to hear from you if there are some difficulties loading it
down or to use it.
please inform me .
WE can do a play:
You send me a single SE(at equilibrium ) scan by E-mail of one of your
scans with
the information of V = the part.spec. volume and the solution density.
BUT PLEASE
DON'T give me the Mw. Later we will compaire with your result from the
same scan.
Thanks to all of you ....ariel
Lustig Ariel
Dept. of Biophysical Chemistry / Analytical Ultracentrifuge Lab.TEL....+61
2672187
Biozentrum der Universität
Basel Fax +61 2672189
email
lustig at ubaclu.unibas.ch
Klingelbergstr. 70
CH 4056 Basel
Switzerland
_______________________________________________
RASMB mailing list
RASMB at rasmb-email.bbri.org
http://rasmb-email.bbri.org/mailman/listinfo/rasmb
More information about the RASMB
mailing list