<div dir="ltr"><p class="MsoNormal"><span lang="EN-US">Dear RASMB
members,</span></p>

<p class="MsoNormal"><span lang="EN-US"> </span></p>

<p class="MsoNormal" style="text-align:justify"><span lang="EN-US">We are having a problem with absorbance optics
on XLI centrifuge(XLI_1 on the images attached). Hopefully you can suggest
something that will help. The problem is
that optical density of any samples (which definitely absorb, including for
example BSA (c = 4mg/mL) or pyrene dye (c = 10µM )) recorded by the AUC is
within the zero  level. Nevertheless, at
much higher concentrations (approximately 10 -20 times higher) the optical
density reaches some level of absorbance (~0.1-0.2 OD). The same solutions
checked by normal UV-Vis spectrophotometer and on another XLI centrifuge
(XLI_2), have adequate values of the absorbance level. We have as well checked
our monochromator   on  XLI_2 and it was working ok. Recorded lamp
intensities (air to air) at 230 nm and 530 nm are at the normal level. If we
perform the absorbance scan of a sample (for example Pyrene dye at c = 10µM)  in the intensity mode, we received the same
level of intensities for both reference and sample chamber! The corresponding
comparison of the intensities is shown on the image attached. Any of your
suggestions about the possible problem and/or the solution will be highly
appreciated.</span></p>

<p class="MsoNormal" style="text-align:justify"><span lang="EN-US"><br></span></p><p class="MsoNormal" style="text-align:justify"><span lang="EN-US">Best regards</span></p>

<p class="MsoNormal" style="text-align:justify"><span lang="EN-US">Igor Perevyazko</span></p><p class="MsoNormal" style="text-align:justify"><span lang="EN-US"><br style="font-size:12.8000001907349px;text-align:start"><span style="font-size:12.8000001907349px;text-align:start">PhD Igor Perevyazko</span><br style="font-size:12.8000001907349px;text-align:start"><span style="font-size:12.8000001907349px;text-align:start">Physics Department, Polymer Physics</span><br style="font-size:12.8000001907349px;text-align:start"><span style="font-size:12.8000001907349px;text-align:start">St.Petersburg State University</span><br style="font-size:12.8000001907349px;text-align:start"><span style="font-size:12.8000001907349px;text-align:start">Ul. Ulyanovskaya 1, Petrodvorets</span><br style="font-size:12.8000001907349px;text-align:start"><span style="font-size:12.8000001907349px;text-align:start">St.Petersburg, Russia, 198504</span><br style="font-size:12.8000001907349px;text-align:start"><span style="font-size:12.8000001907349px;text-align:start">Tel.: (812)4284382</span><br style="font-size:12.8000001907349px;text-align:start"><span style="font-size:12.8000001907349px;text-align:start">Fax.: (812)4287240</span><br></span></p></div>