size=2 face=Arial>Frank, I haven't actually done that type of protocol, but in 
principle it should work (assuming the protein is stable enough). My comment 
though is that if the association is weaker at higher temperatures you want to 
do the higher temperature <U>first</U> and then the lower (for the same reason 
you always want to go up in speed for a multi-speed SE sequence). It takes much 
longer to reach equilibrium when the change (temperature or speed) will 
produce a gradient that is less steep than when the gradient 
increases.</FONT></SPAN></DIV>
<DIV dir=ltr align=left><SPAN class=370283215-25102013><FONT color=#0000ff 
size=2 face=Arial></FONT></SPAN> </DIV>
<DIV dir=ltr align=left><SPAN class=370283215-25102013><FONT color=#0000ff 
size=2 face=Arial>John</FONT></SPAN></DIV>
<DIV lang=en-us class=OutlookMessageHeader dir=ltr align=left>
<HR tabIndex=-1>
<FONT size=2 face=Tahoma><B>From:</B> rasmb-bounces@list.rasmb.org 
[mailto:rasmb-bounces@list.rasmb.org] <B>On Behalf Of </B>Frank 
Niesen<BR><B>Sent:</B> Friday, October 25, 2013 7:47 AM<BR><B>To:</B> 
rasmb@rasmb.org<BR><B>Subject:</B> [RASMB] varying temperature during an 
equilibrium run<BR></FONT><BR></DIV>
<DIV></DIV>
<DIV dir=ltr>Hello,
<DIV><BR></DIV>
<DIV>I was wondering if anybody has experience with changing the temperature in 
an equilibrium run at given speed, to investigate temperature-dependent strength 
in association? Is it feasible/sane to record data at equilibrium at, e.g., 20C, 
and then to change the temperature to, e.g., 40C, and to hope to get data for 
equilibrium and record once more?</DIV>
<DIV><BR></DIV>
<DIV>In my case, the association seems to weaken substantially with temperature 
and I observe a weaker curvature at the higher temperature. For a third step, 
going back to the original temperature, I programmed the run, but made two 
mistakes with it - firstly to not allow sufficient time for the equilibrium and, 
secondly, to somehow got the "stop AUC after last scan" checked :o(</DIV>
<DIV>As result, I am looking at a scan that doesn't resemble the one at lower 
temperature, yet it is similar to scans I saw during the equilibration between 
the two temperatures at step 2.</DIV>
<DIV><BR></DIV>
<DIV>Has anybody tried something similar? Am I onto something believable/real, 
or shouldn't waste any more precious sample on a proper repeat (i.e. separate 
runs/samples for each temperature)?</DIV>
<DIV><BR></DIV>
<DIV>Thanks in advance!</DIV>
<DIV>Frank</DIV></DIV></BODY></HTML>