    <div class="moz-cite-prefix">Richard,<br>
      <br>
      in principle, a differential viscometer can yield not only
      relative, but also absolute dynamic viscosities. Both p_i, the
      inlet pressure, and delta p, the pressure difference between the
      solution and the solvent branch, are absolute values - the
      question is, does the instrument software give you access to these
      raw data. The maths I used, however, only worked out for symmetric
      branches, and the instrument we used at that time, a WGE eta-1002,
      had assymmetrical branches. I never got to the point of solving
      the more complicated case. Eventually, the viscometer broke down,
      and I abandoned the project.<br>
      <br>
      However, a differential viscometer intended for use as an SEC
      detector is very sensitive, and our measurements gave good data
      also for diluted solutions. Consider that we had a constant eluent
      flow and injected as little as 100 µl into that flow. So the
      signal was actually for a solution yet more diluted than the bulk
      solution. The intriguing point about this project was that if you
      couple a second detector into the flow, like a UV, for
      concentration of the solute, you get two sets of data: absolute
      viscosities and absolute concentrations for every data point of
      the elugram. This is equivalent to a number of independent
      measurements at varied concentration. Thus, it should in principle
      be possible to calculate even an intrinsic viscosity from a single
      shot.<br>
      <br>
      The drawback is that the eluent flow is split into the two
      branches, where a reservoir on the solvent branch, filled with
      solvent at the beginning of the experiment, gradually fills with
      solution containing the solute. You have only a couple of shots
      until the solvent branch is contaminated with solute, and the
      system needs to be purged. <br>
      <br>
      Though I never brought this to an end, it still appears a
      promising option that should work out - in principle.<br>
      <br>
      Kind regards,<br>
      Kristian<br>
      <br>
      Am 07.06.2013 05:53, schrieb Richard Kingston:<br>
    </div>
    <blockquote
      cite="mid:8D622599-D89D-48F6-B61F-E01F79090A9A@auckland.ac.nz"
      type="cite">
      <meta http-equiv="Content-Type" content="text/html;
        charset=ISO-8859-1">
      <div><br>
      </div>
      <div><br>
      </div>
      Thanks for the many informative replies, both on- and off- list.
      <div><br>
      </div>
      <div>The consensus seems to be that differential
        (pressure-imbalance) viscometers like the Viscostar are well
        suited to the measurement of protein intrinsic viscosities, and
        there are several in commercial production. However these
        instruments, by design, measure only relative viscosities </div>
      <div><br>
      </div>
      <div>Reframing my original question, I was really wondering if it
        were possible to find an instrument which could measure absolute
        solution viscosities with sufficient accuracy to allow protein
        characterisation (and could therefore do double duty, being able
        to measure solution viscosity for other biophysical
        applications).  However for the reasons outlined by Arthur, and
        as attested below by Dave, it might be difficult to achieve good
        results this way.</div>
      <div><br>
      </div>
      <div>Best,</div>
      <div><br>
      </div>
      <div>Richard</div>
      <br>
    </blockquote>
    <br>
    <pre class="moz-signature" cols="72">-- 
Nanolytics
Gesellschaft fuer Kolloidanalytik mbH
Dr. Kristian Schilling

Am Muehlenberg 11
D-14476 Potsdam
Tel: +49 331 5818360
Fax: +49 331 5818361
e-mail: <a class="moz-txt-link-abbreviated" href="mailto:schilling@nanolytics.de">schilling@nanolytics.de</a>
Internet:  <a class="moz-txt-link-abbreviated" href="http://www.nanolytics.de">www.nanolytics.de</a>
_______________________________________________
Diese E-Mail kann vertrauliche und/oder rechtlich geschützte Informationen enthalten. Wenn Sie nicht der richtige Adressat sind oder diese E-Mail irrtümlich erhalten haben, informieren Sie bitte sofort den Absender und vernichten Sie diese Mail. Das unerlaubte Kopieren sowie die unbefugte Weitergabe dieser Mail ist nicht gestattet.

This e-mail may contain confidential and/or privileged information. If you are not the intended recipient (or have received this e-mail in error) please notify the sender immediately and destroy this e-mail. Any unauthorized copying, disclosure or distribution of the material in this e-mail is strictly forbidden.</pre>
  </body>
</html>