<P><FONT color=#0000ff size=2 face=Arial>Matt, these type of problems, and 
things such as motion problems with the slit assembly, are exactly what the 
absorbance log files <U>should</U> be able to help diagnose. But 
the log files generated by the GUI do not record basic information such as 
when each scan starts and stops, and hence they are <U>useless</U> to both 
the users and service technicians. Making the log files useful should be a 
trivial software change, and a number of us have asked for this for at least 10 
years now, but nothing has ever been done.</FONT></P>
<DIV><FONT size=2><FONT color=#0000ff face=Arial>It is also long past time for 
Beckman to clarify exactly what the two time stamps recorded in the headers 
(and also the Windows file time stamp) mean. Are these the values when the 
scan is first initiated (before the instrument goes to 6.5 cm to read the 
intensities and set the PMT voltage and gain), or after that step occurs, or at 
the end of the scan? Again software writers have asked for such clarifications 
many times, but to my knowledge none of us have ever gotten an 
answer.</FONT></DIV>
<P><FONT color=#0000ff face=Arial>John</FONT><BR><BR>-----Original 
Message-----<BR>From: rasmb-bounces@list.rasmb.org [<A 
href="mailto:rasmb-bounces@list.rasmb.org">mailto:rasmb-bounces@list.rasmb.org</A>] 
On Behalf Of Rhyner, Matthew N<BR>Sent: Friday, January 25, 2013 10:48 AM<BR>To: 
John Burgner; rasmb@list.rasmb.org<BR>Cc: 'Walter Stafford'<BR>Subject: Re: 
[RASMB] Time-stamp checks for SV data<BR><BR>All-<BR><BR>Thank you for these 
messages.  We are aware of the issue and are actively investigating the 
root cause.<BR><BR>Matthew N Rhyner, PhD<BR>Strategic Product Manager<BR>Beckman 
Coulter, Inc.<BR>Mobile: 404-313-1811<BR><BR><BR>-----Original 
Message-----<BR>From: rasmb-bounces@list.rasmb.org [<A 
href="mailto:rasmb-bounces@list.rasmb.org">mailto:rasmb-bounces@list.rasmb.org</A>] 
On Behalf Of John Burgner<BR>Sent: Friday, January 25, 2013 1:32 PM<BR>To: 
rasmb@list.rasmb.org<BR>Cc: 'Walter Stafford'<BR>Subject: Re: [RASMB] Time-stamp 
checks for SV data<BR><BR>Taking any large time interval from a series of scans 
and calculating the w^2dt for that interval gives the same value calculated from 
the w^2t values in the file header so yes both values are incorrect.  Note 
also that windows stores file creation times accurate to 1 sec.<BR>John 
Burgner<BR><BR>-----Original Message-----<BR>From: rasmb-bounces@list.rasmb.org 
[<A 
href="mailto:rasmb-bounces@list.rasmb.org">mailto:rasmb-bounces@list.rasmb.org</A>] 
On Behalf Of Walter Stafford<BR>Sent: Friday, January 25, 2013 10:38 AM<BR>To: 
Borries Demeler<BR>Cc: rasmb@list.rasmb.org<BR>Subject: Re: [RASMB] Time-stamp 
checks for SV 
data<BR><BR>Borries,<BR><BR>        The time 
values in the file header are "elapsed time" from the the time the rotor starts 
turning and should not be used in analysis in any case. SEDANAL uses 
integral(w^2 dt), which is the correct value to use for all analyses.<BR><BR>So 
your question is important: Are those values reported correctly or not in 
version 6?<BR><BR>Walter<BR><BR>_________________________<BR>Walter 
Stafford<BR>wstafford3@walterstafford.com<BR><BR><BR><BR>_________________________<BR>Walter 
Stafford<BR>wstafford3@walterstafford.com<BR><BR><BR><BR>On Jan 25, 2013, at 
10:13, Borries Demeler wrote:<BR><BR>> Rodolfo:<BR>><BR>> are w^2t 
values reported in the file header not affected, or has that<BR>> not been 
investigated? I do not have this version to test myself, but<BR>> such errors 
would propagate into some analysis methods in UltraScan as<BR>> well. As far 
as I know the w^2t values are reported directly from the<BR>> DA system of 
the XLA, I am not certain where the time values<BR>> originate. The w^2t 
values *may* be affected as well, and that should<BR>> be<BR>looked at by 
someone with a Proteomelab ver.6.<BR>><BR>> 
-borries<BR>><BR>><BR>> On Fri, Jan 25, 2013 at 09:29:39AM -0500, 
Ghirlando, Rodolfo<BR>> (NIH/NIDDK)<BR>[E] wrote:<BR>>> Dear friends 
and colleagues,<BR>>><BR>>> It has come to our (Joy Zhao, Peter 
Schuck, Grzegorz Piszczek, Chad<BR>Brautigam and myself) attention that 
sedimentation coefficients based on SV data collected using the ProteomeLab 
XLA/I version 6 acquisition software are up to 10% larger than expected 
(depending on the rotor speed).<BR>>><BR>>> We have traced this to a 
smaller than expected elapsed time stamp in<BR>>> the<BR>file header. A 
correction for this has been implemented in SEDFIT based in part on the Windows 
OS file creation time stamp. This version has been released, as indicated in the 
forwarded email from the SEDFIT-L list below, and we will report more details on 
this shortly.<BR>>><BR>>> In addition Beckman Coulter has been 
advised of this issue.<BR>>><BR>>> 
Sincerely,<BR>>><BR>>> Rodolfo 
Ghirlando<BR>>><BR>>><BR>>> FORWARDED 
MESSAGE<BR>>><BR>>><BR>>> Dear 
Colleagues,<BR>>><BR>>> This is to let you know of a critical SEDFIT 
update with version<BR>>> 14.0c, which you can download from<BR>>> 
<A 
href="https://sedfitsedphat.nibib.nih.gov/software/Shared%20Documents/sedfi">https://sedfitsedphat.nibib.nih.gov/software/Shared%20Documents/sedfi</A><BR>>> 
t140c.zip<BR>>><BR>>> It has a very important new feature for the 
analysis of sedimentation<BR>velocity: it automatically compares the elapsed 
times reported in the data file headers with the time differences obtained from 
the file creation timestamps generated by the Windows operating system on the 
PC.<BR>>><BR>>> Why would we need this? In the course of a joint 
study between our<BR>>> lab<BR>and the labs of Rodolfo Ghirlando, Greg 
Piszczek and Chad Brautigam, we have recently discovered that data files can 
have elapsed time entries less than the differences from the operating system 
timestamp, by as much as 10%.<BR>Obviously, if the elapsed time entry is 10% too 
low then this leads to s-values that are 10% too high!<BR>>><BR>>> 
We have double checked data from many different instruments, and<BR>correlated 
this discrepancy in elapsed times with the ProteomeLab data acquisition software 
version 6. No time differences, or only trivial ones, were observed for 
instruments running earlier versions of the data acquisition software. In some 
runs the overall difference amounts to hours at the end of the run, and I 
believe that the operating system timestamp is correct. We will distribute more 
detailed information shortly, but want to give you already a heads-up in the 
meantime with this note: If you are running version 6 you may want to examine 
the data files, and/or time the data acquisition process yourself, to 
verify.<BR>>><BR>>> We have discussed this matter with our Beckman 
service engineer and<BR>>> it<BR>appears as though the company is now 
addressing this issue.<BR>>><BR>>> The new SEDFIT release can fix 
this problem for the data analysis.<BR>>> First,<BR>it automatically 
checks and creates a report if a systematic and significant discrepancy exists 
between differences in the file time-stamp and differences in the header elapsed 
time. If there appears to be a problem (i.e. apparent dilation factor greater 
than a user-defined threshold, initially defaulted to 1.005 or 0.5%), SEDFIT 
will create a report and then offer to write data files (under different names, 
of course) with time entries based on a dilation factor extracted from the file 
timestamps.<BR>Subsequent analysis of these time-corrected data files should 
eliminate the problem, as it did perfectly for the data in our study. Such 
time-corrected scan files should also be the ones loaded into SEDPHAT, which 
will get its own update shortly.<BR>>><BR>>> Please let me know if 
you have any questions, or if any problems<BR>>> occur<BR>with the new 
version.<BR>>><BR>>> Best wishes and good luck,<BR>>> 
Peter<BR>>><BR>>><BR>>> Peter Schuck, PhD<BR>>> Chief, 
Dynamics of Macromolecular Assembly Section Laboratory of<BR>>> Cellular 
Imaging and Macromolecular Biophysics National Institute of<BR>>> 
Biomedical Imaging and Bioengineering, NIH<BR>>> 13 South 
Drive<BR>>> Bldg 13, Rm 3N17<BR>>> Bethesda, MD 20892<BR>>> 
phone: (301) 435-1950<BR>>> fax: (301) 480-1242<BR>>> email: 
schuckp@mail.nih.gov<BR>>><BR>>><BR>><BR>>> 
_______________________________________________<BR>>> RASMB mailing 
list<BR>>> RASMB@list.rasmb.org<BR>>> <A 
href="http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org">http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org</A><BR>><BR>> 
_______________________________________________<BR>> RASMB mailing 
list<BR>> RASMB@list.rasmb.org<BR>> <A 
href="http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org">http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org</A><BR><BR>_______________________________________________<BR>RASMB 
mailing list<BR>RASMB@list.rasmb.org<BR><A 
href="http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org">http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org</A><BR><BR>_______________________________________________<BR>RASMB 
mailing list<BR>RASMB@list.rasmb.org<BR><A 
href="http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org">http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org</A><BR><BR><BR>Please 
be advised that this email may contain confidential information.  If you 
are not the intended recipient, please notify us by email by replying to the 
sender and delete this message.  The sender disclaims that the content of 
this email constitutes an offer to enter into, or the acceptance of, any 
agreement; provided that the foregoing does not invalidate the binding effect of 
any digital or other electronic reproduction of a manual signature that is 
included in any 
attachment.<BR><BR><BR>_______________________________________________<BR>RASMB 