<DIV dir=ltr align=left>
<DIV dir=ltr align=left><SPAN class=657433420-10032011><FONT color=#0000ff 
size=2 face=Arial>Ian, this is a problem <SPAN 
class=525024220-10032011>which I believe</SPAN> occurs in all XL-I's. Basically 
under some circumstances the CPU on the data acquisition board (DAB) <SPAN 
class=525024220-10032011>gets overloaded and </SPAN>just can't keep up with 
everything it needs to do. One symptom that this is occurring is indeed that the 
XL Monitor window on the computer runs several minutes behind real-time. 
</FONT></SPAN></DIV>
<DIV dir=ltr align=left><SPAN class=657433420-10032011><FONT color=#0000ff 
size=2 face=Arial></FONT></SPAN> </DIV>
<DIV dir=ltr align=left><SPAN class=657433420-10032011><FONT face=Arial><FONT 
color=#0000ff><FONT size=2>The circumstances that usually create this problem 
are running both optical systems together and <SPAN 
class=525024220-10032011>using </SPAN>high rotor speeds. I actually don't think 
I've <U>ever</U> been able to complete a run at 60K rpm <SPAN 
class=525024220-10032011>when </SPAN>collecting both absorbance and interference 
data.<SPAN class=525024220-10032011> Increasing the interval between scans will 
help avoid crashes.</SPAN></FONT></FONT></FONT></SPAN></DIV>
<DIV dir=ltr align=left><SPAN class=657433420-10032011><FONT color=#0000ff 
size=2 face=Arial></FONT></SPAN> </DIV>
<DIV dir=ltr align=left><SPAN class=657433420-10032011><FONT color=#0000ff 
size=2 face=Arial>Because of this problem between runs you always should cut the 
power to the instrument to make the DAB clear itself out. If it starts off 
'confused' you may hit trouble under circumstances where otherwise you would 
not. Restarting the computer will <U>not</U> help.</FONT></SPAN></DIV>
<DIV dir=ltr align=left><SPAN class=657433420-10032011><FONT color=#0000ff 
size=2 face=Arial></FONT></SPAN> </DIV>
<DIV dir=ltr align=left><SPAN class=657433420-10032011><FONT color=#0000ff 
size=2 face=Arial>John</FONT></SPAN></DIV><BR>
<HR tabIndex=-1>
<FONT size=2 face=Tahoma><B>From:</B> rasmb-bounces@rasmb.bbri.org 
[mailto:rasmb-bounces@rasmb.bbri.org] <B>On Behalf Of </B>Ian 
Berke<BR><B>Sent:</B> Thursday, March 10, 2011 12:23 PM<BR><B>To:</B> 
rasmb@rasmb.bbri.org<BR><B>Cc:</B> david.keller@yale.edu; Incarvito, 
Christopher<BR><B>Subject:</B> [RASMB] CPU Board Malfunction on 
XL-I<BR></FONT><BR></DIV>
<DIV></DIV>Hi all,
<DIV><BR></DIV>
<DIV>We have an XL-I machine (~1.5yr old) which has given a CPU board 
malfunction error (code#11) several times now.  It happens after several 
hours and the centrifuge just shuts down while the computer still thinks it's 
scanning.  Beckman tells us it's a software problem on the computer and an 
upcoming software release should fix it.  They suggest that in the meantime 
I do a short absorbance scan (r 6.2-6.7) of the counterbalance to allow the 
computer's data saving to catch up with the AUC scans of other cells.</DIV>
<DIV><BR></DIV>
<DIV>I don't know if it's related, but a couple of times I also noticed that the 
centrifuge and the computer were out of sync as the centrifuge accelerated to 
speed.  I started the centrifuge from the computer and began calibrating 
the laser for IF.  That part went as normal, but when I looked at the XL 
monitor on the computer, the speed readings were delayed by ~1-2min and the 
computer would be waiting for speed and not letting me start scanning until it 
had caught up.</DIV>
<DIV><BR></DIV>
<DIV>Has anyone else experienced this problem?</DIV>
<DIV><BR></DIV>
<DIV>Thanks,</DIV>
<DIV>Ian</DIV></BODY></HTML>