Greetings All:<br><br>
What is the experience of others for the stray light level at 250
nm?  I know that 3 is the specification, but Is 1.5 abs pushing the
limit at 250 nm? <br><br>
Borries' point about gradients is good, especially given the solubility
issues.  A check of the buffer might help.<br><br>
Glen<br><br>
At 10:50 PM 9/6/2009, Paul Leonard wrote:<br>
<blockquote type=cite class=cite cite="">Dear Borries Demeler,<br><br>
Thank-you for replying so quickly.<br><br>
In response to your suggestions I have a few comments (although
these<br>
responses are not directed directly to you but to the RASMB mailing
list<br>
as a whole - please enjoy the labor day weekend rather than responding
to<br>
me).<br><br>
I collected abs measurements at both 280nm and 250nm.  For the
250nm<br>
curves there was no problem with the absorbance going above 1.5 abs
units<br>
so I used the same radial range as I used with the interference
optics.<br><br>
I believe this rules out aggregation as being responsible for the<br>
difference in curvature between abs and interference.<br><br>
I really do not know how BeF3- would behave in the experiment. 
Actually I<br>
should clarify this a little.  BeF3- is actually the addition of 5.3
mM<br>
BeCl2 and 33 mM NaF.  Previous studies by NMR have found that at
these<br>
concentrations the major beryllium fluoride species present is BeF3-
(and<br>
not BeF2, BeF42- etc).  However BeF2 is relatively insoluble as is
MgF2<br>
which would also be present in small amounts in my sample.<br><br>
However I guess I over complicated the issue (or at least me
greatest<br>
concern) by mentioning the BeF3- experiments.  Even in the
experiments<br>
where BeF3- was not present I still do not get the same answer when<br>
fitting the absorbance and interference optics so there must be
something<br>
fundamentally wrong with what I am doing as the two techniques are
viewing<br>
the exact same protein distribution.<br><br>
<br>
kind regards<br><br>
Paul<br><br>
<br><br>
> Paul,<br>
><br>
> A couple of thoughts: You may have some aggregation in your
equilibrium<br>
> experiment (perhaps because of the longer run time). Perhaps in
the<br>
> absorbance<br>
> experiment you are clipping the higher OD's because they are out
of<br>
> the dynamic range of the optics, but in the interference
experiment<br>
> the dynamic range is larger, and you are including the portion of
the<br>
> scan that correlates most with the aggregates.<br>
><br>
> Another possibility is the presence of time invariant noise, but
that<br>
> should have been subtracted from the data by your baseline
scan.<br>
><br>
> Finally, I wonder if the BeF3 refractive index is properly handled
by<br>
> having a meniscus-matched BeF3 subtracted. I don't know the BeF3
density<br>
> off hand, but dependening on density it could for a gradient
and<br>
> significantly contribute to the refractive index signal.<br>
><br>
> Others may have addtl. ideas...<br>
><br>
> -borries<br>
><br>
>> Dear all,<br>
>><br>
>> I have been running both sedimentation velocity and
sedimentation<br>
>> equilibrium AUC experiments on a relatively simple protein
system.<br>
>><br>
>> The sedimentation velocity experiments at multiple protein<br>
>> concentrations<br>
>> show that the protein is just a monomer unless BeF3- is present
(to<br>
>> mimic<br>
>> phosphorylation) and then all of the protein becomes
dimeric.<br>
>><br>
>> However if I run sedimentation equilibrium experiments at
multiple<br>
>> speeds<br>
>> I find that if I use the absorbance optics it is in agreement
with the<br>
>> sedimentation velocity result - protein is just a monomer in the
absence<br>
>> of BeF3- but fits as a dimer if BeF3- is present.<br>
>><br>
>> However if I look at the interference scans for the same cells
(for<br>
>> sedimentation equilibrium AUC) the protein looks considerably
bigger -<br>
>> i.e. it would appear that it is in equilibrium between monomer
and dimer<br>
>> in the absence of BeF3-<br>
>><br>
>> I have also record analytical size exclusion chromatography
profiles for<br>
>> concentrations of the protein all the way from 10uM upto 500 uM
and if<br>
>> BeF3- is absent I cannot see any evidence of dimer
formation.  If I add<br>
>> BeF3- the protein is all dimer.<br>
>><br>
>> Therefore I am pretty certain that I should only be seeing
monomer in<br>
>> the<br>
>> absence of BeF3- and dimer if BeF3- is present so I'd like to
know why<br>
>> my<br>
>> interference data makes the protein look larger than it
should.  Is<br>
>> there<br>
>> some correction that I need to apply.  I have recorded
water blanks at<br>
>> each speed and it makes virtually no difference if I subtract
these<br>
>> scans<br>
>> (Not enough of a correction to account for the additional
curvature in<br>
>> the<br>
>> scans).<br>
>><br>
>> Could it be that the radial calibration is wrong?  Is
something wrong<br>
>> with<br>
>> the interference optics?  Any explanation would be greatly
appreciated.<br>
>><br>
>> I fear that if we cannot get reliable results for this simple
case using<br>
>> this equipment for more complex systems will become impossible
(at least<br>
>> for sedimentation equilibrium experiments).<br>
>><br>
>> Thank-you for any help you can provide.<br>
>><br>
>> Paul<br>
><br><br>
<br>
Department of Biochemistry<br>
UMDNJ-Robert Wood Johnson Medical School<br>
Center for Advanced Biotechnology and Medicine<br>
679 Hoes Lane<br>
Piscataway, New Jersey 08854-5627<br><br>
Phone: (732) 235-4206<br>
Email: leonard@cabm.rutgers.edu<br><br>
_______________________________________________<br>
RASMB mailing list<br>
RASMB@rasmb.bbri.org<br>
<a href="http://rasmb.bbri.org/mailman/listinfo/rasmb" eudora="autourl">
http://rasmb.bbri.org/mailman/listinfo/rasmb</a></blockquote>
<x-sigsep><p></x-sigsep>
<br>
   Glen Ramsay, Ph.D.<br>
   Chief Scientist<br>
   Aviv Biomedical, Inc.<br>
   750 Vassar Avenue<br>
   Lakewood, NJ 08701<br>
   (732) 370-1300, Ext. 25<br>
   (732) 370-1303, FAX<br>
   glen@avivbiomedical.com<br>
  
<a href="http://www.avivbiomedical.com/" eudora="autourl">
www.avivbiomedical.com<br>
</a>   Skype: GlenAtAviv<br><br>
<font size=1>The information transmitted is intended only for the person
or entity to which it is addressed and may contain confidential and/or
privileged material, the disclosure of which is governed by applicable
law. Any review, retransmission, dissemination or other use of, or taking
of any action in reliance upon, this information by persons or entities
other than the intended recipient is prohibited. If you received this in
error please contact the sender and destroy the materials contained in
this message. </font></body>
</html>