<div>If I allign the cell properly wouldn't I expect the screw hole to orient towards the center properly?</div>
<div>I indeed didn't mention what material I used. I used charcoal filled epon centerpiece. Recommeneded speed 42000rpm but we routinely use it at 50000rpm like many other people. But the speed doesn't solve it as the same cell assmebly that didn't leak with water leaked with the protein sample the very next time.</div>

<div> </div>
<div>Sumit</div>
<div><br><br> </div>
<div class="gmail_quote">On Fri, Jun 12, 2009 at 4:51 PM, sumit goswami <span dir="ltr"><<a href="mailto:ikant11@gmail.com">ikant11@gmail.com</a>></span> wrote:<br>
<blockquote class="gmail_quote" style="PADDING-LEFT: 1ex; MARGIN: 0px 0px 0px 0.8ex; BORDER-LEFT: #ccc 1px solid">
<div>Dear all</div>
<div>            we have been trying do some sed-velocity experiments for last couple of weeks but every time we tried most of the cell assemblies leaked!! Apparently it seemed like a simple problem of aging of center pieces but then very recently I ran all the seven cell assemblies overnight at 50000rpm with water in them and none leaked! After that I used the same set of assembly for running my sample by sucking off the water from them  and 4 out of 5 leaked even at 4000rpm. We have used different kinds of buffer so far... PE (5mM Na-phosphate, 1mM EDTA), TE (10mM Tris, 1mM EDTA), Poly buffer (MES,tris, acetate). All of our buffer had 0.01%PEG in them. We have tried both PH 5 & 8.</div>

<div>   Can anyone suggest what's going on? We are ready to buy new centerpiece but not sure if that's the problem. The result with water is confusing us!</div>
<div> </div>
<div>Thank you very much!</div>
<div> </div>
<div>Regards,</div>
<div>Sumit </div></blockquote></div><br>