<div dir="ltr"><div>Hi all,</div>
<div>I have recently seen a vacuum jump from 9 microns to 45 microns during the acceleration to 50K rpm. At the same time, the temperature goes up to 20.4C from the set temp of 20.0C. It came back stepwise to 20C after 2 minutes based on the temperature plot in sedfit.  My experience is that this cooling took too long. The data fit was really noisy with the rmsd of 0.015 (total OD280=0.9) or above and the residual was bad for the first 45 scans. </div>

<div>Is this a vacuum issue or the temperature control problem? Any input will be appreciated.  Thanks.</div>
<div> </div>
<div> </div>
<div> </div>
<div>Chin</div></div>