<DIV>Here is the lamp intensity check procedure from a 10-2001 manual.  I'd 
follow this to step 8 then just look at it. The origin part is a waste of time 
in my opinion.</DIV>
<DIV>John Gunhter</DIV>
<DIV><BR><FONT face=Arial size=2>*********** REPLY SEPARATOR 
***********<BR><BR>On 4/4/2008 at 10:57 AM John Correia wrote:</FONT></DIV>
<BLOCKQUOTE 
style="PADDING-LEFT: 5px; MARGIN-LEFT: 5px; BORDER-LEFT: #000000 2px solid">
  <DIV><FONT face=Arial size=3>It is worth noting that Beckman has a service 
  manual for the XLA XLI that discusses all of these features and tests and 
  specs.  The spec for linearity is 20% when measured at 280 nm on a 
  windowless cell (or an empty hole) at 3000 rpm using OD values 1 cm 
  apart.  The old manual specified 6.0 and 7.0 cm.  With the old lamp 
  rotation of the lamp strongly influenced the intensity across the cell; the 
  newer lamp holder is now symmetric and do not vary as much with rotation, 
  in my experience.  The old part # for this manual is 679045 but my copy 
  is dated 1/92.  The newest version I am aware of is called Proteome Lab 
  XLA & XLI Service Manual dated (10/3).  Sorry I do not have a part # 
  - the relevant chapter is section 3.  It describes wavelength calibration 
  and accuracy checks, Intensity checks, cleaning the optics, radial position 
  calibration, etc, etc.  Obviously depends upon the software you have as 
  well.   In my opinion it is well worth buying or getting a copy for 
  your service guy - in the long run it saves both of you time.  Why this 
  isn't standard operation procedure with Beckman is a very good 
  question.</FONT></DIV>
  <DIV><FONT face=Arial size=3></FONT> </DIV>
  <DIV><FONT face=Arial size=3> </DIV>
  <DIV> </DIV>
  <DIV> </DIV>
  <DIV>-------------------------------------------------------------------<BR>Dr. 
  John J. "Jack" Correia<BR>Department of Biochemistry<BR>University of 
  Mississippi Medical Center<BR>2500 North State Street<BR>Jackson, MS  
  39216<BR>(601) 
  984-1522                                 
  <BR>fax (601) 
  984-1501                             
  <BR>email address: jcorreia@biochem.umsmed.edu     
  <BR>homepage location: <A 
  href="http://biochemistry.umc.edu/correia.html">http://biochemistry.umc.edu/correia.html</A><BR>dept 
  homepage location:    <A 
  href="http://biochemistry.umc.edu/">http://biochemistry.umc.edu/</A><BR>-------------------------------------------------------------------<BR><BR><BR><BR></FONT><BR>>>> 
  On 4/4/2008 at 10:18 AM, in message 
  <C7B1B01A69FC4164B0213B9168C3CBAA@79B7XD1>, "John Philo" 
  <jphilo@mailway.com> wrote:<BR></DIV>
  <DIV 
  style="PADDING-LEFT: 7px; MARGIN: 0px 0px 0px 15px; BORDER-LEFT: #050505 1px solid; BACKGROUND-COLOR: #f3f3f3">Mitra,<BR><BR>To 
  my knowledge there is no specification for intensity variation across 
  the<BR>cell, but in my experience few instruments would vary as little as 
  10%.<BR>Probably 20-30% is more typical. It is a common mistake to think that 
  the<BR>absolute intensities have a strong effect on the signal/noise. It 
  is<BR>doubtful that you would notice the effect of even a 50% drop in 
  intensity<BR>(assuming you are starting from a normal level). Further, 
  positioning the<BR>lamp to give the maximum possible intensity will not 
  necessarily give the<BR>best overall performance---quite often the maximum 
  intensity position also<BR>gives high stray light and compromises the 
  linearity. <BR><BR>John<BR><BR>-----Original Message-----<BR>From: 
  mitrana@mail.utexas.edu [mailto:mitrana@mail.utexas.edu] <BR>Sent: Thursday, 
  April 03, 2008 6:00 PM<BR>To: jphilo@mailway.com<BR>Cc: 'RASMB'<BR>Subject: 
  RE: [RASMB] Lamp Diagnostics<BR><BR>John<BR>I feel better already.<BR><BR>What 
  I meant in 1. is that I take a wavelength scan at 5.9, 6.5, and 7.1 cm<BR>and 
  compare the 230 nm peaks. I guess a radial scan is definitely a better<BR>way 
  to check intensity variability across a cell. How much variability 
  is<BR>considered acceptable - somehow 10% seems stuck in my 
  head.<BR><BR>Mitra<BR><BR>Quoting John Philo 
  <jphilo@mailway.com>:<BR><BR>> Mitra,<BR>><BR>> Points 2 and 3 
  are normal. Every time the monochromator moves there is <BR>> an 
  uncertainty of +/- 1-2 nm (I believe the official specification is <BR>> 
  actually<BR>> +/- 4 nm).<BR>><BR>> Also for wavelength scans the 
  monochromator doesn't actually take data <BR>> at the exact wavelength 
  interval you ask for. So sometimes it will <BR>> essentially step over the 
  sharp peaks (if you double-click on the line <BR>> graph and ask it to mark 
  the data points with symbols you can more <BR>> easily see there are no 
  points where the peak should be).<BR>><BR>> I'm not sure I understand 
  your point 1, but I think the intensity <BR>> variability may just be a 
  consequence of the wavelength variability <BR>> discussed above. Normally 
  one would try to adjust the lamp alignment <BR>> using radial scans, not 
  wavelength scans, and typically the wavelength <BR>> intensity scans to 
  check the strength of the 230 nm peak or wavelength <BR>> calibration are 
  run at 6.5 cm.<BR>><BR>> John<BR>><BR>> -----Original 
  Message-----<BR>> From: rasmb-bounces@rasmb.bbri.org <BR>> 
  [mailto:rasmb-bounces@rasmb.bbri.org] On Behalf Of <BR>> 
  mitrana@mail.utexas.edu<BR>> Sent: Thursday, April 03, 2008 2:58 PM<BR>> 
  To: 'RASMB'<BR>> Subject: [RASMB] Lamp Diagnostics<BR>><BR>> Hello 
  Spinners<BR>> Our AUC started giving some problems recently. Did a 
  wavelength scan <BR>> and here's a summary of the results -<BR>><BR>> 
  1. Lamp intensity varies at the 5.9 cm position although at the 
  'sweet<BR>spot'<BR>> obtained from the lamp alignment this variation is 
  about 9-10%.<BR>><BR>> 2. Sometimes it seems the peaks are cut-off and I 
  get a plateau <BR>> instead of a peak with of course a lower intensity 
  count.<BR>><BR>> 3. The peak oscillates 1-2 nm (227-230 nm and 526-528 
  nm) on <BR>> successive scans at a single radial position.<BR>><BR>> 
  It seems like a monochromator problem to me but I'd like to hear some <BR>> 
  comments from experts here. I am particularly worried about the peaks <BR>> 
  getting cut-off.<BR>> Our facility did not renew the service contract this 
  year (I know, <BR>> extremely bad idea but too many size exclusion column 
  huggers around) <BR>> and I'd rather know the problem instead of being told 
  by the Beckman <BR>> Service person that it's within specs while charging 
  $290.00 per hour.<BR>><BR>> Mitra<BR>><BR>> --<BR>> Mitra S. 
  Rana<BR>> Graduate Student<BR>> Institute for Cellular and Molecular 
  Biology 2500 Speedway, UT-Austin <BR>> Austin, TX-78712 
  _______________________________________________<BR>> RASMB mailing 
  list<BR>> RASMB@rasmb.bbri.org<BR>> <A 
  href="http://rasmb.bbri.org/mailman/listinfo/rasmb">http://rasmb.bbri.org/mailman/listinfo/rasmb</A><BR>><BR><BR><BR>--<BR>Mitra 
  S. Rana<BR>Graduate Student<BR>Institute for Cellular and Molecular Biology 
  2500 Speedway, UT-Austin<BR>Austin, 
  TX-78712<BR><BR>_______________________________________________<BR>RASMB 
  mailing list<BR>RASMB@rasmb.bbri.org<BR><A 
  href="http://rasmb.bbri.org/mailman/listinfo/rasmb">http://rasmb.bbri.org/mailman/listinfo/rasmb</A><BR><BR></DIV><BR 
  clear=all>Individuals who have received this information in error or are not 
  authorized to receive it must promptly return or dispose of the information 
  and notify the sender. Those individuals are hereby notified that they are 
  strictly prohibited from reviewing, forwarding, printing, copying, 
  distributing or using this information in any way. <FONT size=2 
Arial></BLOCKQUOTE></FONT></BODY></HTML>