<font size="+2"><tt>I have checked the RASMB archives and there does
not seem to be anything on this particular subject.<br>
The problem is not the XL-I; the problem is an overly complex sample
and reference.<br>
<br>
The sample contains a protein that absorbs at 290 nm as well as a
ligand that absorbs at 290 nm.<br>
The reference contains just the ligand. Both are in matching buffers.<br>
<br>
The first scan: The baseline absorbance starts below zero and becomes
more negative at higher radius.<br>
Scan 100: The sample has almost sedimented completely giving the
baseline absorbance what appears to be a "bow". <br>
Scan 200: The sample has sedimented completely and the absorbance is
well below zero. It is more negative at high radius than low radius.<br>
<br>
The problem is clearly the sample. I have not run the "control" in
which the sample compartment contains no protein but does contain
ligand; reference compartment contains no ligand.<br>
<br>
This problem must have come up in the past. Needless to say, all
analyses are feasible but they are not likely to be correct. I'm at
wits' end.<br>
<br>
thanks for any help you can give. i am attaching an xlgraph of the
three scans.<br>
jack kornblatt<br>