size=2>John,</FONT></SPAN></DIV>
<DIV><SPAN class=166484219-01062007><FONT face=Arial color=#0000ff 
size=2></FONT></SPAN> </DIV>
<DIV><SPAN class=166484219-01062007><FONT face=Arial color=#0000ff size=2>First, 
can you quantify what you mean by the noise is "quite bad". What is the r.m.s. 
noise for a scan of an empty hole (<U>not</U> an empty cell) in the rotor 
now, and what was it in the past?</FONT></SPAN></DIV>
<DIV><SPAN class=166484219-01062007><FONT face=Arial color=#0000ff 
size=2></FONT></SPAN> </DIV>
<DIV><SPAN class=166484219-01062007><FONT face=Arial color=#0000ff size=2>These 
intensity readings are never absolute---20,000 counts on one instrument might be 
10,000 on one and 30,000 on another. The only thing that counts is what is 
the ratio of the intensity when the lamp is clean versus now. Even a 2-fold 
drop of intensity usually has a minimal impact on the noise level, so the real 
source of the problem may be something else altogether (perhaps a dirty slit 
assembly). </FONT></SPAN></DIV>
<DIV><SPAN class=166484219-01062007><FONT face=Arial color=#0000ff 
size=2></FONT></SPAN> </DIV>
<DIV><SPAN class=166484219-01062007><FONT face=Arial color=#0000ff 
size=2>John</FONT></SPAN></DIV>
<BLOCKQUOTE style="MARGIN-RIGHT: 0px">
  <DIV></DIV>
  <DIV class=OutlookMessageHeader lang=en-us dir=ltr align=left><FONT 
  face=Tahoma size=2>-----Original Message-----<BR><B>From:</B> 
  rasmb-bounces@rasmb.bbri.org [mailto:rasmb-bounces@rasmb.bbri.org] <B>On 
  Behalf Of </B>John E Harlan<BR><B>Sent:</B> Friday, June 01, 2007 11:49 
  AM<BR><B>To:</B> rasmb@server1.bbri.org<BR><B>Subject:</B> [RASMB] Lamp 
  Intensities on Analytical Ultracentrifuges<BR><BR></FONT></DIV><BR><FONT 
  face=Arial size=2>I am starting to have some problems with rapidly decreasing 
  lamp intensities on my analytical instruments.  Within about 1 month of 
  moderate use after my service people were in (they cleaned the optics while 
  checking other things), I find the noise in my velocity absorbance scans is 
  quite bad.   When I check the intensities I find maxima at 230nm to be 
  8300 for an XLA and 13000 for an XLI, well below the 20000 to 30000 I should 
  find for a well functioning instrument.</FONT> <BR><BR><FONT face=Arial 
  size=2>I have two questions:</FONT> <BR><BR><FONT face=Arial size=2>1) 
   What could cause such a rapid decay of performance.  I suspect oil 
  leaks fouling the optics (I have observed droplets in the can on at least one 
  occasion).  I understand alignment can also cause problems.  Is 
  there something else to ask my service people to check?</FONT> <BR><BR><FONT 
  face=Arial size=2>2)  I have heard that people that have upgraded to a 
  turbo-molecular vacuum pump have been quite pleased with performance. 
   Can this upgrade help keep my optics clean, if indeed that is the 
  problem?</FONT> <BR><BR><FONT face=Arial size=2>Thanks for your help in 
  advance.</FONT> <BR><BR><FONT face=Arial 
size=2>John</FONT></BLOCKQUOTE></BODY></HTML>