<br><font size=2 face="Arial">I am starting to have some problems with
rapidly decreasing lamp intensities on my analytical instruments.  Within
about 1 month of moderate use after my service people were in (they cleaned
the optics while checking other things), I find the noise in my velocity
absorbance scans is quite bad.   When I check the intensities I find
maxima at 230nm to be 8300 for an XLA and 13000 for an XLI, well below
the 20000 to 30000 I should find for a well functioning instrument.</font>
<br>
<br><font size=2 face="Arial">I have two questions:</font>
<br>
<br><font size=2 face="Arial">1)  What could cause such a rapid decay
of performance.  I suspect oil leaks fouling the optics (I have observed
droplets in the can on at least one occasion).  I understand alignment
can also cause problems.  Is there something else to ask my service
people to check?</font>
<br>
<br><font size=2 face="Arial">2)  I have heard that people that have
upgraded to a turbo-molecular vacuum pump have been quite pleased with
performance.  Can this upgrade help keep my optics clean, if indeed
that is the problem?</font>
<br>
<br><font size=2 face="Arial">Thanks for your help in advance.</font>
<br>
<br><font size=2 face="Arial">John</font>