<DIV><SPAN class=838072621-24032006><FONT face=Arial color=#0000ff size=2>It is 
hard to comment without knowing more about the experiment or seeing the data. 
Can you at least tell us whether you are talking about absorbance or 
interference data? </FONT></SPAN></DIV>
<DIV><SPAN class=838072621-24032006><FONT face=Arial color=#0000ff 
size=2></FONT></SPAN> </DIV>
<DIV><SPAN class=838072621-24032006><FONT face=Arial color=#0000ff size=2>You 
mention that you have considered non-ideality, which suggests that this could be 
an experiment at high protein concentration. Is that correct? If so, then part 
of the problem might be optical distortions due to refraction by the 
boundary---these are always worst early in the run, and go away as diffusion 
makes the gradient less steep. That problem can be significantly reduced by 
using 3 mm centerpieces.</FONT></SPAN></DIV>
<DIV><SPAN class=838072621-24032006><FONT face=Arial color=#0000ff 
size=2></FONT></SPAN> </DIV>
<DIV><SPAN class=838072621-24032006><FONT face=Arial color=#0000ff 
size=2>John</FONT></SPAN></DIV></DIV>
<BLOCKQUOTE dir=ltr style="MARGIN-RIGHT: 0px">
  <DIV></DIV>
  <DIV class=OutlookMessageHeader lang=en-us dir=ltr align=left><FONT 
  face=Tahoma size=2>-----Original Message-----<BR><B>From:</B> 
  rasmb-bounces@rasmb.bbri.org [mailto:rasmb-bounces@rasmb.bbri.org] <B>On 
  Behalf Of </B>David Hayes<BR><B>Sent:</B> Friday, March 24, 2006 12:41 
  PM<BR><B>To:</B> RASMB<BR><B>Subject:</B> [RASMB] Sedimentation Velocity 
  Experiments that are hard to fit<BR><BR></FONT></DIV>Hi to all experienced 
  Ultracentrifuge Scientists,<BR><BR>I have some data (repeated experiments) 
  from sample A buffer 1 (which may or may not have been produced on the planet 
  earth -- that is all I can say about it)<BR><BR>There are two puzzling aspects 
  to this data.<BR>First, I am not getting good repeatability in the amount of 
  aggregate.<BR>Second, by eye, the raw data looks like a single species (which 
  it should be) with a small amount of higher weight aggregate (not surprising), 
  but the entire data set never fits well no matter what I try.<BR>Thinking that 
  the repeatability problem might be an artifact of my inexperience using Sedfit 
  I tried many different strategies:  including floating or not floating 
  just about every parameter and also using the nifty "experimental initial 
  distribution" option of Sedfit.<BR>  Fitting to the whole data set, I got 
  a fit with a single large peak and two very small aggregate bumps which all 
  together integrate to about 4% of the total.  The rmsd was a bit large at 
  0.01 and the residuals were systematic.  Early scans were fit very poorly 
  and later scans fit somewhat poorly with scans in the middle fit pretty 
  well.  Later using Sedfit with the single species model and only the 
  later half of the scans and floating a lot of things and turning off the hat 
  function in Sedfit brought me down to a comparable fit with a rmsd of .006, 
  but the residuals were still systematic and the molecular weight was off more 
  than what I expected for having 4% aggregate.<BR>Then I did something that 
  Peter Shuck thinks is a bit strange, but I think it helped me understand the 
  data a bit more.  I fit pairs of scans with c(s) taking scans at 
  different times paired with the last scan.  Essentially then, I was 
  fitting each scan individually:  using the last scan in ever pair was 
  needed to help Sedfit know the baseline.  Peter reminded me that by 
  fitting each scan, I lose the ability to distinguish heterogeneity from 
  diffusion.  But in this case, all the evidence points to a single 
  species, so I did not think heterogeneity (beyond the measured 4% aggregation) 
  was significant.  The peak S value stays constant all the way down the 
  cell.  Each individual scan fit this way fit very well without systematic 
  residuals and a rmsd of about .005.  What changed was that early scans 
  fit to an f/f0 of 1.06 while by the later scans f/f0 fit to 1.5.  This 
  means that the early scans are broader than they should be for a reasonable 
  f/f0 and that the later scans seem to be narrower and diffusing less.  
  <BR>I repeated this with Sedanal dcdt based curve fitting (here I took scans 
  by twos going down the cell and fit to a single species ignoring the 4% 
  aggregate) and found a similar trend:  the fitted molecular weight went 
  from 95,000 to 225,000 and then settled down to about 160,000.  If there 
  were heterogeneity, the scans would fit the opposite way to a one species 
  model, the f/f0 would go down, or the weight would go up later in the run as 
  the species sedimented apart.<BR><BR>I had some interference data of 
  Tropomyosin that I subjected to a similar type of analysis with Sedanal, 
  fitting the whole data set and then fitting scans by sets of 10 for the 
  different times in the experiment (because of TI, RI noise fitting, it is not 
  possible to fit small sets of scans with Sedfit using interference data; 
  however, a Sedfit fit to the whole Tm data set showed no systematic variation 
  of the residuals dependent on scan number).  There was no scan number 
  dependent trend in molecular weight with this molecule, though the variability 
  in fitted molecular weight was surprising (weights from different sets of 10 
  varied from 60,000 to 80,000 randomly).<BR><BR>Seeing this I tried a Sedfit 
  single species and Sedanal single species fits with non-ideality parameters, 
  but floating these parameters made no substantial difference.<BR><BR>I don't 
  know if anyone has run into this type of problem, but I am out of ideas what 
  to try next.  I am being a bit stubborn wanting a good fit to the 
  data:  the S value of the main peak is reproducible and the data contains 
  quite a bit of information, but I can't figure out exactly what is going 
  on:  <BR><BR>Is this just non-ideality that I am not fitting properly or 
  maybe that is not modeled well?<BR>Is the culprit convection?  Thoughts 
  on convection in the next message.<BR><BR>Cheers<BR><BR><BR><X-SIGSEP>
  <P></X-SIGSEP>Dr. David B Hayes<BR>Boston Biomedical Research Institute<BR>64 
  Grove St.<BR>Watertown, MA  02472<BR>617-658-7738<BR><BR>
  <DIV align=center><B>Boston Biomedical Research Institute... Today's Research 
  for Tomorrow's Health.<X-TAB> </X-TAB><BR><I>Please visit us at <A 
  href="http://www.bbri.org/" eudora="autourl">www.bbri.org</A></B></I> 
</DIV></BLOCKQUOTE></BODY></HTML>