If the standards are being ran at the same flow rate, and have the same
peak widths as the Ribonulcease A, then I think you can exclude hardware
issues.  If the hardware were at fault then its "scan
rate" would have to slow down, or your elution rate sped
up.<br><br>
Try a different separation methodology to detect multiple species. 
Ion exchange with the pH close to the "pI"  might separate
small charge differences that might result from proteolytic cleavage.
  <br><br>
Glen<br><br>
At 08:22 PM 2/3/2006, Richard Kingston wrote:<br><br>
<blockquote type=cite class=cite cite="">Recently, when running molecular
weight standards over a size  <br>
exclusion column I made a puzzling observation.<br><br>
I routinely monitor  column outflow at three wavelengths, since
our  <br>
chromatography system allows it (I'm using an Amersham Biosciences 
<br>
Äkta). This is to ensure that an undistorted elution profile is 
<br>
obtained at at least one wavelength (For concentrated, and highly 
<br>
absorbing samples, the detector response may become non-linear). So 
<br>
typically I would monitor the signal at 280 nm (the peak due to Trp/
Tyr/Cys) and at then on the shoulder of this peak (293 and 300nm)
for  <br>
reduced sensitivity. Or when increased sensitivity is required, I 
<br>
often monitor at 220 or 230 nm, on the far shoulder of the peptide 
<br>
bond absorbance peak.<br><br>
In the past,  when using the instrument in this way, traces
have  <br>
always been identical (up to some scaling factor) when the
absorbance  <br>
is < 1.0.<br><br>
But last week, when running standards over a Superdex 75 column,
one  <br>
of them (Ribonuclease A) gave different apparent elution volumes at 
<br>
different wavelengths (280, 293 and 300nm). The behavior was  <br>
reproducible, and the shifts in the peak position were small yet 
<br>
significant (far beyond the inherent uncertainty in the peak  <br>
position). Apparent elution volume at 280 nm > 293 nm > 300 nm.
This  <br>
also corresponds to the order in which the wavelengths are cycled
by  <br>
the monochomator.<br><br>
Yet the remaining standards (Chymotrypsinogen A, Ovalbumin,
Albumin,  <br>
Blue Dextran 2000, L-Tryptophan) , run before and afterwards, in an 
<br>
identical fashion, did not exhibit this behavior. In these cases
the  <br>
traces at different wavelengths were all equivalent, accounting for 
<br>
the differing absorbance by the sample. These standards are both 
<br>
smaller and larger than Ribonuclease A.<br><br>
So the behavior is both puzzling and concerning (what wavelength 
<br>
should I 'believe'  in the case of Ribonuclease A?) and I'm
left  <br>
scratching for an explanation. I'm hoping somebody on the list
might  <br>
suggest something.<br><br>
The UV Monitor on the Akta has a 1 cm path length continuous flow 
<br>
cell. Light is provided by a xenon flash lamp,  monochromated with
a  <br>
grating, and there is a beam splitter which sends 50% pf the light 
<br>
through the cell, and the remainder through a reference fiber. Up
to  <br>
three wavelengths can be recorded 'simultaneously'  (in practice
the  <br>
wavelength is rapidly switched, with the reported switch time < 500
ms).<br><br>
Thanks for any suggestions<br><br>
Rich.<br><br>
Richard Kingston, PhD<br>
School of Biological Sciences<br>
The University of Auckland<br>
Auckland<br>
New Zealand.<br><br>
rl.kingston@auckland.ac.nz<br><br>
<br>
_______________________________________________<br>
RASMB mailing list<br>
RASMB@rasmb.bbri.org<br>
<a href="http://rasmb.bbri.org/mailman/listinfo/rasmb" eudora="autourl">
http://rasmb.bbri.org/mailman/listinfo/rasmb</a></blockquote>
<x-sigsep><p></x-sigsep>
   Glen Ramsay, Ph.D.<br>
   Chief Scientist<br>
   Aviv Biomedical<br>
   750 Vassar Avenue<br>
   Lakewood, NJ 08701<br>
   (732) 370-1300, Ext. 25<br>
   (732) 370-1303, FAX<br>
   glen@avivbiomedical.com<br>
  
<a href="http://www.avivbiomedical.com/" eudora="autourl">
www.avivbiomedical.com<br><br>
</a><font size=1>The information transmitted is intended only for the
person or entity to which it is addressed and may contain confidential
and/or privileged material, the disclosure of which is governed by
applicable law. Any review, retransmission, dissemination or other use
of, or taking of any action in reliance upon, this information by persons
or entities other than the intended recipient is prohibited. If you
received this in error please contact the sender and destroy the
materials contained in this message. </font></body>
</html>