<DIV><FONT size=2>I seriously doubt that you will find a much better way to 
calibrate temperature than the CoCl2 approach Walter published.  Trying to 
use solubility introduces factors like concentration & nucleation effects 
into the mix.  It also only allows a measurement at one 
temperature.</FONT></DIV>
<DIV><FONT size=2></FONT> </DIV>
<DIV><FONT size=2>At the intro & service Beckman courses in Palo Alto at the 
end of one of the days they half-fill two sectors of a prep rotor with vacuum or 
drive oil, spin overnight, & put a thermometer into the solutions to check 
temperature.  Its one point per day but I guess a good way to check even 
for prep centrifuges.</FONT></DIV>
<DIV><FONT size=2></FONT> </DIV>
<DIV><FONT size=2>I certainly think this standardization is a good idea, & I 
am glad to participate, although I must stress again I calibrate temperature and 
we measure density, and density*vbar is a much larger source of error.  So 
get a density meter while you're at it!</FONT></DIV>
<DIV><FONT size=2></FONT> </DIV>
<DIV><FONT size=2> I would also encourage you to recognize this is not the 
experts coming to the rescue.  Any physical measurement requires 
calibration and careful regulation of temperature.  This is a fact of life, 
not a special event that will somehow resolve the problem for everybody & 
all machines.  You must wait for temperature before hitting start, & if 
you want to do careful work you must calibrate.  (I asked for a poll 
earlier - if you have calibrated temperature on your XLA/I send me a 
note!)</FONT></DIV>
<DIV><FONT size=2></FONT> </DIV>
<DIV><FONT size=2>The suggestions to do replicates is not something unique to 
this event.  You must always do replicates, as a function of concentration, 
to understand the behavior of your system.  The sample Arthur sends to us 
will presumably be nonassociating, low charge, homogeneous, & in a buffer 
near the isoelectric point.  For typical samples association, 
heterogeneity or nonideal behavior are the sources of deviation.  In 
my lab multiple users typically achieve 0.1 - 0.3% reproducible S values on well 
behaved noninteracting systems, & this applies to joint analysis with 
sedfit, DCDT+ and SEDANAL.  (For monomers the analysis method should not be 
that critical - but it is the holidays, so get in the spirit & sell 
away.)  I also honestly think users doing their own analysis should be part 
of the experiment, for no reason other than setting a standard & expecting 
the field to meet that standard.  I think I have now crossed the PC line - 
</FONT></DIV>
<DIV><FONT size=2></FONT> </DIV>
<DIV><FONT size=2>happy holidays</FONT></DIV>
<DIV><FONT size=2></FONT> </DIV>
<DIV> </DIV>
<DIV><FONT 
size=2>-------------------------------------------------------------------<BR> Dr. 
John J. "Jack" Correia<BR> Department of Biochemistry<BR> University 
of Mississippi Medical Center<BR> 2500 North State Street<BR> Jackson, 
MS  39216<BR> (601) 
984-1522                                 
<BR> fax (601) 
984-1501                             
<BR> email address: <A 
href="mailto:jcorreia@biochem.umsmed.edu">jcorreia@biochem.umsmed.edu</A>     
<BR> homepage location: <A 
href="http://biochemistry.umc.edu/correia.html">http://biochemistry.umc.edu/correia.html</A><BR> dept 
homepage location:    <A 
href="http://biochemistry.umc.edu/">http://biochemistry.umc.edu/</A><BR>-------------------------------------------------------------------<BR> <BR> <BR></FONT></DIV></BODY></HTML>