<DIV><FONT size=2>I suspect other will give technical responses about the 
ability of these temperature controllers to function at the +/- 0.1 C 
level.  My take is slightly different.  </FONT></DIV>
<DIV><FONT size=2></FONT> </DIV>
<DIV><FONT size=2>What matters is not the set point but knowing the actual 
temperature.</FONT></DIV>
<DIV><FONT size=2></FONT> </DIV>
<DIV><FONT size=2>To my knowledge most users never use the Stafford & LUS 
method to calibrate the temperature on their machine.  (can we vote on the 
RASMB in some way?) It is no more tedious than waiting three hours before 
starting a run.  I have checked the calibration on my machine twice, once 
in '93 when I got it, and again in 1999 when a result caused me to question 
the temperature accuracy.  A series of measurements up to 40 C and 
back down to 4 C (with parallel measurements in a spec) takes about two 
days.  It does not take three hours to go from 20 to 25 C, as verified by 
the stability of the area under the CoCl2/ethanol curve.  At a setting of 4 
C I get 3.6, at 20 I get 19.7, at 40 I get 39.8.  The values drifted by 0.1 
C in 7 years.  So if we mean by accuracy +/- 0.5 C of setting, my machine 
is within spec.  When I fit data I use the actual temperature, not the set 
point. & I trust the setting on the screen at low vacuum and hit start when 
it reports the set temperature.</FONT></DIV>
<DIV><FONT size=2></FONT> </DIV>
<DIV><FONT size=2>Comparisons between uncalibrated machines honestly make no 
sense to me.  Calibrate & be done with it.  (Arthur, if you 
do that in Nottingham & still get different values of S between machines I 
would be concerned, & amazed.)  </FONT></DIV>
<DIV><FONT size=2></FONT> </DIV>
<DIV><FONT size=2>I also use the calibrated temperature values when I measure 
density in an Anton Paar DMA 5000.  It has a peltier cell good between 0 - 
80 C, so I dial in 19.7 instead of 20 C and measure away.  I suspect the 
error from a calculated density (viscosity & Vbar) is larger than 
(albeit coupled to) an assumed temperature.  </FONT></DIV>
<DIV><FONT size=2></FONT> </DIV>
<DIV><FONT size=2>At this point the inquiring "student" should assume 
errors in various parameters and propagate them into S or MW by the appropriate 
equations - my favorite book for teaching this is Bevington, "Data Reduction and 
Error Analysis for the Physical Sciences".</FONT></DIV>
<DIV><FONT size=2></FONT> </DIV>
<DIV><FONT size=2>PS - why Beckman has never joined up with Anton Paar to bundle 
Density Meters into XLA/XLI quotes amazes me.</FONT></DIV>
<DIV><FONT size=2></FONT> </DIV>
<DIV><FONT size=2>PPSS - Walter claims to keep the original solution 
around, in a cell, from the CoCl2 calibration work, and years later it still 
gives the same results.  Quick checking may not be as difficult as one 
might think.</FONT></DIV>
<DIV> </DIV>
<DIV> </DIV>
<DIV> </DIV>
<DIV>-------------------------------------------------------------------<BR> Dr. 
John J. "Jack" Correia<BR> Department of Biochemistry<BR> University 
of Mississippi Medical Center<BR> 2500 North State Street<BR> Jackson, 
MS  39216<BR> (601) 
984-1522                                 
<BR> fax (601) 
984-1501                             
<BR> email address: <A 
href="mailto:jcorreia@biochem.umsmed.edu">jcorreia@biochem.umsmed.edu</A>     
<BR> homepage location: <A 
href="http://biochemistry.umc.edu/correia.html">http://biochemistry.umc.edu/correia.html</A><BR> dept 
homepage location:    <A 
href="http://biochemistry.umc.edu/">http://biochemistry.umc.edu/</A><BR>-------------------------------------------------------------------<BR> <BR> <BR><BR><BR>>>> 
Arthur Rowe <arthur.rowe@nottingham.ac.uk> 12/10/04 08:17AM 
>>><BR>Hi Everyone <FONT color=#000080>{this is a second (</FONT><FONT 
color=#ff00ff>now 3rd!</FONT><FONT color=#000080>) try at getting this mail out 
- first attempt got lost in cyber-space, it seems}<BR></FONT><BR>Mei-Ling Chien 
gives us a very useful review of the nature of the temperature measurement and 
control system in the XL-I/A instrument. However, I do not think that this fully 
addresses the problems which one has in determining what the absolute 
temperature of one's sample actually <I>is</I> when it is going round in the 
rotor at speed.<BR><BR>It is, of course, only a worry to those (very limited) 
number of people for whom an absolute s value is of importance, normally for 
hydrodynamic modelling purposes (although formulation issues should not be 
forgotten). When I raised this issue on RASMB a week or so back, my concern was 
not  "<FONT color=#ff0000><TT>to ensure their operation within the 
published specification".</TT></FONT>    . I am trying to get the 
<U>accuracy</U> of the temperature read-out to be close to the <U>precision</U> 
of which the system is capable. I have no evidence at all to suggest that the 
accuracy is outside the quoted spec of 0.5º. It is just that I - in my greedy 
way - want 0.1º.<BR><BR>Even the method mentioned (equilibrate for 3 hours - 
under vacuum - and then check  "<FONT color=#ff0000><TT>with a calibrated 
external temperature sensing device to verify accuracy"</TT></FONT> 
  is not unambiguous in what it will yield. Quite apart from matters 
such as adiabatic effects when one releases the vacuum to use an "external 
temperature sensing device", can one be sure that the thermal emissivity of a 
spinning rotor surface, averaged over everything that is passing by, is equal to 
that of a piece of the rotor surface 'seen' in a stationary rotor? <BR><BR>None 
of these are new concerns, and I certainly lay no claim to the IPRs! I imagine, 
from what Mei-Ling Chien has communicated, that we at least know clearly that 
the ±0.5º refers to the accuracy of the temperature <U>as measured by the 
defined procedure</U>. Walter Stafford's colorimetric method (Stafford & 
Liu) did not suggest the presence of errors outside the stated accuracy limit, 
and is surely a valid way to approach the absolute temperature issue. But is is 
pretty tedious to use as a procedure, and certainly as a routine QA method is 
not feasible. <BR><BR><FONT color=#008000><I>As an approach to the size of the 
problem, would there be support for Borries Demeler's suggestion (a single 
sample to be circulated and multiple users on multiple machines to report an s 
value under defined conditions)? After all, the NCMH + Borries's Lab gives us 6 
machines for starters.<BR></I></FONT><BR>Any way, we here keep trying here to 
locate the holy grail - a simple, cheap, effective method for determining the 
in-cell temperature to ±0.1º <BR><BR>Regards to all (and many thanks to Mei-Ling 
Chien)<BR><BR>Arthur<BR><BR>-- <BR>*************************<BR>Arthur 
Rowe<BR>Lab at Sutton Bonington<BR>tel: +44 115 951 6156<BR>fax: +44 115 951 
6157<BR>*************************<BR></DIV>
<BLOCKQUOTE><BR><B>From: </B>mchien@beckman.com<BR><B>Date: </B>Fri, 3 Dec 
  2004 10:11:46 -0800<BR><B>To: </B>"'rasmb@rasmb-email.bbri.org'" 
  <rasmb@server1.bbri.org><BR><B>Subject: </B>[RASMB] Re: XL-A/I temp 
  control<BR><BR></BLOCKQUOTE><BR>
<BLOCKQUOTE><TT>----------------------------------------------------------------------------------<BR>The 
  older archived RASMB emails can be found 
  at:<BR>http://rasmb-email.bbri.org/rasmb_archives<BR>and current archives 
  at<BR>http://rasmb-email.bbri.org/pipermail/rasmb/<BR>Search All the Archives 
  at:<BR>http://rasmb-email.bbri.org/rasmb_search.html<BR>----------------------------------------------------------------------------------<BR><BR>Hi 
  All,<BR><BR>Below is response regarding XL-A/I temperature control from our 
  Technical<BR>Support 
  Department.<BR><BR>******************************************************<BR>Mei-Ling 
  Chien PhD<BR>Staff Development Scientist, Centrifugation<BR>Platform & 
  Automation Business Center<BR>Beckman Coulter 
  Inc.<BR><BR>mchien@beckman.com<BR>(650) 
  859-1948<BR>******************************************************<BR><BR><BR>The 
  basis for temperature control specifications were instrument 
  design<BR>specifications for temperature control and dynamic system testing 
  during the<BR>prototype phase of the product.<BR><BR>If there is a discrepancy 
  in temperature control and measurement between<BR>instruments of the same 
  design then a dynamic calibration check should be<BR>performed on both 
  instruments <FONT color=#ff0000>to ensure their operation within the 
  published<BR>specification.<BR></FONT><BR>First the physical condition of 
  components within the temperature control and<BR>vacuum system should be 
  verified through inspection.  Then an electronic<BR>calibration for 
  temperature control and vacuum can be performed.  Lastly a<BR>dynamic 
  test or rotor dunk test is performed (rotor should be precooled 
  or<BR>preheated to avoid testing delay).  The rotor and its contents must 
  be allowed<BR>to equilibrate for up to 3 hours or more.  When set 
  temperature equals indicated<BR>temperature at the instrument interface, the 
  rotor temperature is then checked<BR><FONT color=#ff0000>with a calibrated 
  external temperature sensing device to verify accuracy</FONT>.<BR><BR>If the 
  checks fall out of specification then appropriate troubleshooting 
  is<BR>required to isolate the electronic or mechanical fault in the 
  temperature<BR>control or vacuum system.  Once the fault is corrected the 
  temperature control<BR>checks are performed again.<BR><BR>Quote from Bob 
  Giebeler, Analytical Ultracentrifugation in Biochemistry<BR>  and 
  Polymer Science, 1992,16-25 for the Optima XLA/I.<BR>"Temperature 
   control is considerably more stable, provides more rapid 
  cool-down<BR>and  heat-up  rates,  is  thermally more 
  uniform, and has equivalent accuracy as<BR>compared  to previous models 
  including the Model E.  This control system uses an<BR>isothermal 
  radiometer temperature-sensing system to sense the temperature of the<BR>rotor 
   that  is  emissivity-independent  ad  view 
   factor-corrected in software.<BR>Heating  and cooling of the rotor 
  are accomplished by the refrigeration can that<BR>surrounds  the 
   rotor,  which  is  in  turn  heated and cooled 
  by thermoelectric<BR>modules.   This environment is very isothermal, 
  and at equilibrium, irrespective<BR>of  speed  or  temperature, 
   rotor temperature is within about one degree of the<BR>refrigeration can 
  temperature.<BR><BR><BR>The  control  system  that 
   regulates  rotor  temperature,  as  monitored by 
  the<BR>radiometer,  is  highly  software-intensive. 
    This  software encompasses 
  triple<BR>proportional-integral-differential  control algorithms and 
  proportional-integral<BR>smoothing  algorithms.  In addition, 
  radiometer view factors are measured during<BR>rotor   cool-down 
   to  allow  more  rapid  rotor  cool-down 
   and  more  accurate<BR>temperature  monitoring 
   during  cool-down.  While at equilibrium, refrigeration<BR>can 
    temperature   fluctuation   does  not 
   typically  exceed  +0.5C,  and  the<BR>corresponding 
  rotor temperature fluctuation is less than 
  +0.2C.<BR><BR><BR><BR><BR><BR><BR><BR><BR><BR><BR><BR>_______________________________________________<BR>RASMB 
  mailing 
  list<BR>RASMB@rasmb-email.bbri.org<BR>http://rasmb-email.bbri.org/mailman/listinfo/rasmb<BR></TT></BLOCKQUOTE><TT><BR></TT><BR>
<P>This message has been scanned but we cannot guarantee that it and any 
attachments are free from viruses or other damaging content: you are advised to 
perform your own checks. Email communications with the University of Nottingham 
may be monitored as permitted by UK legislation. </P></BODY></HTML>