size=2>Samantha,</FONT></SPAN></DIV>
<DIV><SPAN class=575363719-23012004><FONT face=Arial color=#0000ff 
size=2></FONT></SPAN> </DIV>
<DIV><SPAN class=575363719-23012004><FONT face=Arial color=#0000ff size=2>It is 
very common with the <EM>c(s)</EM> method to see what you are calling a 
'half-peak' at the left end, but this is often a 'false' peak. Basically it 
arises because the data do not go out to long enough times to provide any real 
information about the very small sedimentation coefficients, because they never 
pull significantly away from the meniscus. If you are analyzing data taken at 
30K rpm, you probably really can't 'see' any species below 0.5 S (if that). If 
you allow the fitter to play with parameters that have no real influence on the 
data, it is virtually guaranteed that it will come up with some nonsense trying 
to fit what is really noise in the data.</FONT></SPAN></DIV>
<DIV><SPAN class=575363719-23012004><FONT face=Arial color=#0000ff 
size=2></FONT></SPAN> </DIV>
<DIV><SPAN class=575363719-23012004><FONT face=Arial color=#0000ff size=2>So you 
probably want to start the distribution at a higher s value, but doing that 
doesn't always get rid of your half-peak. Once you realize it is false, just 
ignore it. If you need a graph with that peak gone, just copy and paste the 
distribution table into another graphing program and delete those low 
sedimentation coefficient points.</FONT></SPAN></DIV>
<DIV><SPAN class=575363719-23012004><FONT face=Arial color=#0000ff 
size=2></FONT></SPAN> </DIV>
<DIV><SPAN class=575363719-23012004><FONT face=Arial color=#0000ff size=2>It is 
of course possible with interference data that you are actually detecting some 
sedimentation of buffer components, but that is much less likely at 30K. As Walt 
Stafford often points out, even if you use a dialysate the buffer sedimentation 
will not perfectly cancel unless the reference meniscus is matched to the sample 
meniscus.</FONT></SPAN></DIV>
<DIV><SPAN class=575363719-23012004><FONT face=Arial color=#0000ff 
size=2></FONT></SPAN> </DIV>
<DIV><SPAN class=575363719-23012004><FONT face=Arial color=#0000ff size=2>John 
Philo</FONT></SPAN></DIV>
<BLOCKQUOTE dir=ltr style="MARGIN-RIGHT: 0px">
  <DIV></DIV>
  <DIV class=OutlookMessageHeader lang=en-us dir=ltr align=left><FONT 
  face=Tahoma size=2>-----Original Message-----<BR><B>From:</B> 
  rasmb-admin@rasmb-email.bbri.org [mailto:rasmb-admin@rasmb-email.bbri.org] 
  <B>On Behalf Of </B>Jones, Samantha<BR><B>Sent:</B> Friday, January 23, 2004 
  9:20 AM<BR><B>To:</B> rasmb@rasmb-email.bbri.org<BR><B>Subject:</B> [RASMB] 
  Low s value peaks<BR><BR></FONT></DIV>
  <DIV><SPAN class=459145716-23012004><FONT face=Arial size=2>Hello 
  all</FONT></SPAN></DIV>
  <DIV><FONT face=Arial><FONT size=2><SPAN class=459145716-23012004>I have been 
  trying to analyse some interference data on Sedfit and have been repeatedly 
  seeing a</SPAN> <SPAN class=459145716-23012004>large 'half' peak at very 
  low s values on my s value distributions.</SPAN></FONT></FONT></DIV>
  <DIV><FONT face=Arial><FONT size=2><SPAN class=459145716-23012004>On the 
  continuous c(s) distributions after fitting to range of 0.1s to 10s the 
  distribution line starts at c(s) values of 1.5 to 0.5 on 0 on the x axis, 
  dropping to 0 on the y axis by 0.1s value. This is still the case after 
  changing the parameters to 0.01s to 10s. </SPAN></FONT></FONT></DIV>
  <DIV><FONT face=Arial><FONT size=2><SPAN class=459145716-23012004>The buffer 
  used at first was 10mM NaAcetate, 0.2% Azide and 2mM DTT pH 4. At first I 
  thought it could be the DTT but I saw the same thing using a sample in a 
  buffer 10mM Tris 10mM Na Acetate, 0.2% Azide no DTT pH 8. 
  </SPAN></FONT></FONT></DIV>
  <DIV><FONT face=Arial><FONT size=2><SPAN class=459145716-23012004>I have seen 
  this spinning at 30K and 50K for about 16hours.</SPAN></FONT></FONT></DIV>
  <DIV><FONT face=Arial><FONT size=2><SPAN 
  class=459145716-23012004></SPAN></FONT></FONT> </DIV>
  <DIV><FONT face=Arial><FONT size=2><SPAN class=459145716-23012004>Could Azide 
  be the problem? I don't expect either sample to be contaminated, however 
  anything of this size would have been missed by the SDS-PAGE gel run on both 
  samples.</SPAN></FONT></FONT></DIV>
  <DIV><FONT face=Arial><FONT size=2><SPAN 
  class=459145716-23012004></SPAN></FONT></FONT> </DIV>
  <DIV><FONT face=Arial><FONT size=2><SPAN class=459145716-23012004>Thank 
  you</SPAN></FONT></FONT></DIV>
  <DIV><FONT face=Arial><FONT size=2><SPAN 
  class=459145716-23012004></SPAN></FONT></FONT> </DIV>
  <DIV><FONT face=Arial><FONT size=2><SPAN class=459145716-23012004>Samantha 
  Jones</SPAN></FONT></FONT></DIV>
  <DIV><FONT face=Arial><FONT size=2><SPAN class=459145716-23012004><!-- Converted from text/rtf format -->
  <P><SPAN lang=en-gb><I><FONT face="Palatino Linotype" size=2>Samantha 
  Jones</FONT></I></SPAN> <BR><SPAN lang=en-gb><I><FONT face="Palatino Linotype" 
  size=2>MRC Prion Unit</FONT></I></SPAN> <BR><SPAN lang=en-gb><I><FONT 
  face="Palatino Linotype" size=2>St Marys Hospital</FONT></I></SPAN> <BR><SPAN 
  lang=en-gb><I><FONT face="Palatino Linotype" size=2>Norfolk 
  Place</FONT></I></SPAN> <BR><SPAN lang=en-gb><I><FONT face="Palatino Linotype" 
  size=2>London</FONT></I></SPAN> <BR><SPAN lang=en-gb><I><FONT 
  face="Palatino Linotype" size=2>W2 1PG</FONT></I></SPAN> 
  </P></SPAN></FONT></FONT></DIV></BLOCKQUOTE></BODY></HTML>