[RASMB] dn/dc
Zou, Qin
Qin.Zou at pfizer.com
Fri Mar 18 10:21:45 PDT 2016
Not sure if the M.W. from SEC-MALS are larger or smaller than that from mass spec. it would be good to check if the retention time does not shift as a function of loading. If so, it may be due to some self-association on the column that could give a MW other than a monomer.
Qin
Pfizer Inc
-----Original Message-----
From: RASMB [mailto:rasmb-bounces at list.rasmb.org] On Behalf Of Stephen McLaughlin
Sent: Friday, March 18, 2016 4:56 AM
To: Mattia Rocco; Andrew Leech
Cc: rasmb at rasmb.org
Subject: Re: [RASMB] dn/dc
Dear Mattia, Andrew and John,
The BSA standard recovery is as expected. Hence, given that the Zn ions are not going to effect the dn/dc significantly, I think you are all right in suggesting that the error will be a compound of inaccuracy in dn/dc calculation coupled with lower light scattering for a small protein giving rise to an larger error that would be expected when analysing much larger proteins.
Many thanks
Stephen
On 17 Mar 2016, at 16:08, Mattia Rocco wrote:
> I agree with John regarding the dn/dc, but maybe the absolute calibration of the instrument could also be an issue... or the on-line concentration determination... as Andrew Leech pointed out this morning, do you recover the approximately correct molecular weight for, say, BSA?
>
> Mattia
>
> ________________________________
>
> From: RASMB on behalf of John Philo
> Sent: Thu 3/17/2016 4:22 PM
> To: 'Stephen McLaughlin'; rasmb at rasmb.org
> Subject: Re: [RASMB] dn/dc
>
>
>
> Stephen, I think it is unlikely the dn/dc for your protein differs by more than 5% from your calculation. If there is a "major" discrepancy then calibration is not the issue.
>
> John
>
>
>
> -----Original Message-----
> From: RASMB [mailto:rasmb-bounces at list.rasmb.org] On Behalf Of Stephen McLaughlin
> Sent: Thursday, March 17, 2016 5:08 AM
> To: Mattia Rocco <mattia.rocco at hsanmartino.it>
> Cc: rasmb at rasmb.org
> Subject: Re: [RASMB] dn/dc
>
>
>
> Dear Mattia,
>
>
>
> at the minute, it would be good to know if the dn/dc increases or decreases and roughy how much. There is a significant discrepancy between the molar mass measured by SEC-MALS and the expected mass from mass-spec.
>
>
>
> Best
>
>
>
> Stephen
>
>
>
>
>
>
>
> On 17 Mar 2016, at 09:53, Mattia Rocco wrote:
>
>
>
>> Dear Stephen,
>
>>
>
>> it likely will. But calculating a dn/dc is not that reliable anyway... and what's the level of precision you need, after all?
>
>>
>
>> Best - Mattia
>
>>
>
>> ________________________________
>
>>
>
>> From: RASMB on behalf of Stephen McLaughlin
>
>> Sent: Thu 3/17/2016 10:40 AM
>
>> To: rasmb at rasmb.org <mailto:rasmb at rasmb.org>
>
>> Subject: [RASMB] dn/dc
>
>>
>
>>
>
>> Dear all,
>
>>
>
>> I have a small three fingered Zn binding protein. Will the presence of Zn ions affect the refractive increment that can be caluculated from sequence?
>
>>
>
>>
>
>> Best
>
>>
>
>> Stephen
>
>>
>
>>
>
>> ______________________________
>
>>
>
>> Stephen McLaughlin, Ph.D.
>
>>
>
>> Biophysics
>
>> MRC Laboratory of Molecular Biology
>
>> Francis Crick Avenue
>
>> Cambridge Biomedical Campus
>
>> Cambridge, CB2 0QH
>
>>
>
>> Tel.: +44 (0) 1223 267891 (office)
>
>>
>
>> Email:stephenm at mrc-lmb.cam.ac.uk
>
>> _____________________________
>
>>
>
>>
>
>>
>
>>
>
>>
>
>>
>
>>
>
>
>
> _______________________________________________
>
> RASMB mailing list
>
> RASMB at list.rasmb.org <mailto:RASMB at list.rasmb.org>
>
> http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org <http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org>
>
_______________________________________________
RASMB mailing list
RASMB at list.rasmb.org
http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org
More information about the RASMB
mailing list