[RASMB] XLI absorbance problem
Borries Demeler
demeler at biochem.uthscsa.edu
Fri Feb 20 06:31:19 PST 2015
Dear Igor,
I agree with Andrew, this looks like a timing problem, obviously when
you intensity scan an empty hole it doesn't have this problem (no septum,
cell components in the light path) so the timing being a little bit off
still lets it hit the hole, just not a 2.5 degree channel. Can you still
perform an intensity scan when you put an empty cell in the hole? I am
not sure what the radial position was. If the radial calibration is off
you may not be at the position you think you are, so radially scan an empty
cell or the counterbalance and make sure the edges are at reasonable positions.
-b.
On Fri, Feb 20, 2015 at 04:20:53PM +0300, Igor Perevyazko wrote:
> Dear RASMB members,
>
>
>
> We are having a problem with absorbance optics on XLI centrifuge(XLI_1 on
> the images attached). Hopefully you can suggest something that will
> help. The problem is that optical density of any samples (which definitely
> absorb, including for example BSA (c = 4mg/mL) or pyrene dye (c = 10µM ))
> recorded by the AUC is within the zero level. Nevertheless, at much higher
> concentrations (approximately 10 -20 times higher) the optical density
> reaches some level of absorbance (~0.1-0.2 OD). The same solutions checked
> by normal UV-Vis spectrophotometer and on another XLI centrifuge (XLI_2),
> have adequate values of the absorbance level. We have as well checked our
> monochromator on XLI_2 and it was working ok. Recorded lamp intensities
> (air to air) at 230 nm and 530 nm are at the normal level. If we perform
> the absorbance scan of a sample (for example Pyrene dye at c = 10µM) in
> the intensity mode, we received the same level of intensities for both
> reference and sample chamber! The corresponding comparison of the
> intensities is shown on the image attached. Any of your suggestions about
> the possible problem and/or the solution will be highly appreciated.
>
>
> Best regards
>
> Igor Perevyazko
>
>
> PhD Igor Perevyazko
> Physics Department, Polymer Physics
> St.Petersburg State University
> Ul. Ulyanovskaya 1, Petrodvorets
> St.Petersburg, Russia, 198504
> Tel.: (812)4284382
> Fax.: (812)4287240
> _______________________________________________
> RASMB mailing list
> RASMB at list.rasmb.org
> http://list.rasmb.org/listinfo.cgi/rasmb-rasmb.org
More information about the RASMB
mailing list